Hi
I did a directivity test today.
I did this by normalising the Spec An with the coupler in-circuit and with an open DUT port.
The output of the TG went to a 10dB attenuator. This is directly connected to the coupler input. The reference is a 50ohm termination. The DUT is a U.FL adapter (open circuit load).
The coupler output is connected to the Spec An input.
I normalised the Spec An.
I then replaced the DUT with a 50ohm termination.
Now the Spec An shows the difference between a 50ohm DUT load with near zero reflection and a open circuit DUT load with near 100% reflection. This shows the couplers ability to discriminate between a matched and unmatched load.
This test was done with the coupler enclosed in the Hammond case and with ferrites inside to suppress cavity resonance. Your results may vary.
The plot show no indication of any resonance or external RF interference. There are no anomalous spikes or dips. Within the limits of sensitivity, the measurements can be regarded as reliable.
The plot shows quite good sensitivity (for a low budget device) at the lower frequency end. At 350MHz, the sensitivity is -30dB. The -20dB point is at 1.77GHz which roughly defines the limit of usefulness. You definitely wouldn't want to be doing measurements at 3GHz where the sensitivity falls to just -6db.
This e-bay coupler is no substitute for a certified lab grade device but $/dB is still good. It is a usable piece of kit.
I set up a log period antenna with a bandwidth of 400MHz to 1000MHz. I left the 10dB attenuator on the coupler input for consistency. The result is the second attachment.
The claimed bandwidth is slightly wider than the measured bandwidth. The plot looks believable up to 3.2GHz but study of the previous tests will show that above 2GHz, the results are unreliable.