I successfully performed factory calibration multiple times on my TDS3032B, after extending it to four channels
![Cool 8)](https://www.eevblog.com/forum/Smileys/default/cool.gif)
The fast rise time signal was provided by the negative fast rise output of a PG506, so I didn't have to roll anything myself
![Cheesy :D](https://www.eevblog.com/forum/Smileys/default/cheesy.gif)
All required DC voltages were delivered by Fluke 5101B calibrator. For the 80MHz I could use a E4437 signal generator.
For the remaining 55kHz I initially used a function generator, but later I switched to using the output of the I/Q modulator ARB in the E4437.
In general the procedure is not critical, however, I observed the software occasionally proceeded to the next step without me pressing any key, causing the step to fail. Fortunately you can easily repeat failed steps.
All in all you can perform factory calibration in less than one hour, provided you have everything prepared before starting.
Regards