HP 5328B update here:
So, I’ve been going through the maintenance checks on the unit. When the function switch is set in the “check” position, the readout is 10 MHZ, across all the resolution settings. Pretty solid there. Crystal oven warms up nicely.
It’s clear we have an issue though.
Check #1 – “Time Base Stability & Output”
Epic fail. All I get is 00.000 on the display. It doesn’t read the 1MHZ signal being applied from the sig gen. It would seem the timebase is not so stable.
Just for grins, I hooked up a decent, battery powered 10 MHZ signal to the external oscillator BNC on the back and switched it over. No change. No input signals register.
Check #2 – “Self Check”
All good with this one. The unit responded as expected and described in the manual.
Check #3 – “Frequency response & Sensitivity”
Nada’. No readout except 00.000 during this procedure either.
Nope, she’s not reading anything from the outside world it would seem.
The documentation I have been able to get is not that great for this ‘gal, but I’ve got some schematic study ahead of me for sure. i'll post up with any developments as they occur along the way.
Since this Bodger is, “skiing a slope he has no business being on”
I always appreciate any tips, thoughts or suggestions about the project. -JRH