In regard to disconnecting the secondaries with reference to post #32 the measurements were taken with all the secondaries disconnected (well one from each coloured pair)
I've assumed you disconnected each secondary one by one, reconnecting the last one before disconnecting the next, which would be different than
with all the secondaries disconnected. But still that should have been good enough to find which supply was pulling the rest down.
The amount of power to pull all the secondaries down to only 60% should be more power than the whole scope uses,
something should be getting very hot in theory.
"Should I have disconnected those too? Will that be not giving me correct results? Should I disconnect the Z four and measure them all again?"Looking at the schem.
I think it would be safe (for the scope!) to disconnect the 0V AND 37V ends of the 0-6-37 winding (
rd and
ye), but if the other supplies are still connected these disconnected
rd and
ye ends would still be at -1200V because of the still connected
vi and
bl.
"Also could I trouble you for a concise step by step instruction on how to test the Z board four once I have unsoldered them."You could disconnect the 0-6-37 's mid
vi and
bl as well, then you'd just have a 6.3VAC winding and a 31VAC winding that are isolated and not at -1200V, but the solder tags the 0-6-97 were disconnected from would still be at -1200V.
I'd be happy to measure these 6.3VAC and 31VAC at -1200V, but I'd connect the DVM before switching on and not touch my toolbox DVM while it was at -1200V.
I still think the best, safest and easiest idea is to have
all the secondaries disconnected to see what the transformer does fully unloaded, and there'd be no HV then other than the mains.
Also resistance of Primary Winding 40.2ohms
A ThickPhilM's resistance measurement of his primary might be useful.