Author Topic: Chinese Electronic Load 500W module Sousim 500W  (Read 6741 times)
Offline kawalTopic starter
« on: February 15, 2021, 04:56:04 pm »
I decided to take a risk and get one of these RS485 controlled electric loads. Haven't tested so far - waiting for a rs485 dongle .
Spec are supposed to be 500W , 200V ,  60A ,CV, CC, CW and  CR modes.
While build quality seems OK  there is a slight problem  that the system uses recycled power Fets . Fets seem original IRFP264 devices rated @ 250V .
(legs are covered with shiny solder to a point , looks crusty close to case ).
The design is very simple - basically a STM32 used to control 8 transistors via op amps.
Read back of voltage and current is done by STM as well. The built in ADC/DAC are 12 bit  so appropriate for such design from the 80c .
There is 1 thermal sensors (transistor used as diod ) and room for 2 more sensors that would connect in parallel to the one populated.
 Fan is controlled from the ULN2003 - most likely PWN  controlled.
The following users thanked this post: sorin, thm_w
Offline samofab
  • Country: si
Re: Chinese Electronic Load 500W module Sousim 500W
« Reply #1 on: February 15, 2021, 06:44:34 pm »
I have similar unit. I plan to control it using ESP32, but I didn't come around to it yet. Works nicely using the dongle, though.
I'm attaching a python script if you're interested.. (on second thought, I'm not sure if this is the correct one, but I'll leave it here until proven otherwise)
« Last Edit: February 15, 2021, 06:46:39 pm by samofab »
« Reply #2 on: February 15, 2021, 08:13:05 pm »
Script seems to have errors . But thanks anyway.
« Reply #3 on: February 15, 2021, 09:07:24 pm »
Tried connecting to the TX and RX port available on the board.
Looks like I am able to get MODBUS RTU data
« Reply #4 on: February 16, 2021, 12:29:01 am »
Well tested the unit from 1.5V to 31V  and up to 5A  load
the read back was fairly accurate with a slight offset on the current readings of about 10mA .
 I was testing the different  data  groups on Modbus.
The unit  uses
Holding Registers for settings
HR2 Enable/ Disable
HR3 Overvoltage (20000 default =200.00V)
HR4 Amp setting in  mA
HR9 Over current setting (5500= 55A)
Input registers  for readouts
IR6 Voltage Readout
IR7 Current readout
IR8 Power readout
IR9 Resistance setting readout
The fan was not spinning until the unit got hot 45..50C   and the speed did change with temperature. So there is some kind of fan control
I lost communication while trying to change register HR5 and was unable to get it back. Hope  it comes back to life.
Offline ass20
« Reply #5 on: February 22, 2021, 06:00:43 am »
Hi. Are you buy this load from sousim taobao shop? You can found more software for this load  on page of this shop.  Main io protocol is modbus rtu. You can use any modbus rtu tools .. register definition avaliable ...if you do not have baidu account , please say. I can  replace to other cloud.
What is stm32 used on this load ? Thank you
« Reply #6 on: March 12, 2021, 08:29:38 pm »
Yes got the load from sousim taobao
I do not have a  baidu account so I was not able to get any of the software.
The processor is STM 32F100C8 with 12bit ADC and 12bit DAC . Possible to get more bits out of 12 bit adc via oversampling/averaging.  Vendor claims 16 bit resolution.
« Last Edit: March 12, 2021, 08:41:35 pm by kawal »
Offline vasinap
  • Country: cz
« Reply #7 on: May 10, 2021, 06:30:15 pm »
Dear Sir,
did You bring the load back to life? I faced the same and got lost |O.
Hlp pls.
Petr
Offline syau
  • Country: hk
« Reply #8 on: May 11, 2021, 12:03:07 am »
Software as per attached.
The following users thanked this post: kawal
Offline syau
  • Country: hk
« Reply #9 on: May 11, 2021, 11:46:33 pm »
Some more software downloaded from the link in taobao. I haven’t test them nor do any virus scan. Use it at your own risk.
« Reply #10 on: June 04, 2021, 08:40:06 pm »
I have not tried to get the unit working.  Maybe requires a new processor not sure? I have a working unit if someone knows how to extract the firmware I can try.
Offline mechanicalhand
  • Country: ru
« Reply #11 on: July 22, 2021, 06:53:05 pm »
May be wrong  modbus address,  see this attachments !
Offline ass20
« Reply #12 on: October 16, 2021, 10:18:49 pm »
if dc load  do not answer from rs-485 you can reset to default uart baud rate.
reset sequience
short 2 pins in CON3 connector see picture
and on picture i mark swd pins fw may be do not locked....
« Last Edit: October 16, 2021, 10:20:36 pm by ass20 »
Offline MrDean
  • Country: ba
« Reply #13 on: November 26, 2022, 04:20:19 pm »
I have problem with setting values for measurements. Only first (M1) work correct. If select some other points the I get a message that software not received data from power load after setting value. I tried with few softwares and works same. Don't get any message for other points....
Does anyone had the same problem?
Is the documentation and description of the registers available in English?
Thanks
Offline ass20
« Reply #14 on: November 27, 2022, 09:21:46 am »
Thanks
You can translate 通讯协议.xlsx file  from modbus.rar
for test you can use modbuspoll softaware. its work with mbp and mbw files from  modbus.rar
Offline killingtime
« Reply #15 on: May 01, 2023, 06:49:16 pm »
I pushed the xls through Google Translate. The translation is a bit rough but it's better than nothing. Attached in case it's of use to anyone.
Offline Atom
« Reply #16 on: May 06, 2023, 06:21:44 pm »
thank you, i have the 150w version of this and at first it worked really well now i dont know why it works,(using modbus poll) but i cant seem to get data back, so for examle i cant see any of 30000 registers, previously i could read for example the measured current and voltage. now they are all zero.
40000 registers work as intended, i can set for example currents in CC mode and the load will pull the specified current from the PSU, but registers on the pc interface will still stay zero no matter what i do .
what could have happened? did i fry the adc on the stm32 ? did anyone had the same problem?
Offline ass20
« Reply #17 on: March 25, 2024, 08:02:16 am »
new software link from sousim
https://pan.baidu.com/s/1S3htO5PCUhFOIF1yrdYMMA
Password: 1111
I have 200W module version 1.0
but i do not know what is voltage  and current limits...
 Any body have sousim contacts?
« Last Edit: March 25, 2024, 10:22:28 am by ass20 »
Offline tomoliva
« Reply #18 on: June 30, 2024, 06:56:09 pm »
I tried at Baudu but an account required - anybody out there with an account ?
Latest version is dated 25th May 2024- so looks like it's a product that's being worked on....
Offline ass20
« Reply #19 on: July 23, 2024, 04:10:41 pm »
tomoliva
use baidu cloud app for accset to baidu
For
anybody can download firmware from stm32 ? i have old version of ВС load  but pulse and list mode  do not work..
i add modbus.rar from sousim cloud (file date 02.07.2024)
« Last Edit: July 23, 2024, 04:54:19 pm by ass20 »
  • Country: ru
« Reply #20 on: August 12, 2024, 08:36:43 pm »
Dumps from my 500w electronic load modules
Offline Circlotron
« Reply #21 on: August 12, 2024, 09:05:02 pm »
I like how “QC PASS” is silk screened onto the board before it is even assembled, let alone tested.  ::)
The following users thanked this post: jjoonathan
Offline H.O
« Reply #22 on: August 13, 2024, 07:32:19 am »
What's the problem with that?
If it doesn't pass they simply grab a black marker and cross over that [QC PASS]. THEN they'll ship it. ;D
The following users thanked this post: jasonquin
Offline ass20
« Reply #23 on: August 13, 2024, 05:36:33 pm »
dump form old version 200W and pcb name Practical version 1.0.0
Offline ass20
« Reply #24 on: September 23, 2024, 08:05:02 am »
Hi
Please download dumps of stm32 cpu  wirh stlink
Cpu is not locked
Please !!!