If you have Agilent/HP bench power supplies, open them up and check/replace your RIFA X2/Y caps!
While I was trouble-shooting a random problem with my 6643A system power supply (delivered in late 2007), I decided to also inspect the RIFA caps. While the four larger caps on A1 main board were fine, I noticed small cracks on three of the small Y2 caps on the A1 main board and the GPIB board.
Knowing the risk associated with the RIFA caps, it led me to a quest of opening up all my bench power supplies to check the status of the RIFA caps.
Of the three Agilent 3615A I ordered in 2012 for their silent operation, 2 of them showed nearly invisible cracks on the X2 caps. It is very hard to see, it only shows when the light is hitting it the right way.
Of my first and oldest supply, it is a HP branded 6632A dated to 1993, there were visible cracks on all 5 of the RIFA caps.
I ended up replacing all the RIFA caps on each supply and recalibrated each of them to spec, so I can have a common base line across them. I also recapped all electrolytic caps on the 6632A with highest grade Nichicons due to its age. Amazingly, none of the electrolytic cap I replaced showed any sign of bulging or leakage, and every one was right-on in spec after checking for capacitance, D, and ESR after 26+ years. I pulled up the spec of the larger Sprague caps made in the US, they were rated for 12000 hours lifespan. It is a testament to the quality of the HP construction, so I am leaving the non-RIFA caps alone in all my Agilent branded power supplies.
So my lesson learned, replace all older RIFA caps for sure, check the status of newer ones. For electrolytic caps, I don't think there is reason to change them based on time alone unless I am seeing other issues, as the ripple and accuracy for every supply are all well within spec.