Siglent SDM3055 bench meter relay question
« on: June 16, 2019, 06:57:27 am »
Hello,
I have a Siglent SDM3055. Overall I'm very impressed with the meter. However, for the first time I was playing with the dual display mode. The dual display seems to work fine, however when I run dual AC and DC readings I hear the meter relay "clicking" about every second or so. Is that normal?
 

Offline tautech

  • Super Contributor
  • ***
  • Posts: 29523
  • Country: nz
  • Taupaki Technologies Ltd. Siglent Distributor NZ.
    • Taupaki Technologies Ltd.
Re: Siglent SDM3055 bench meter relay question
« Reply #1 on: June 16, 2019, 07:10:46 am »
Hello,
I have a Siglent SDM3055. Overall I'm very impressed with the meter. However, for the first time I was playing with the dual display mode. The dual display seems to work fine, however when I run dual AC and DC readings I hear the meter relay "clicking" about every second or so. Is that normal?
Yep, quite normal.
The relay switches the low level signals after the dividers/shunts etc so contact life isn’t an issue.
Avid Rabid Hobbyist.