So if you add the characterization process into it then yes, you need the calibrator and other stuff but, at least according to Fluke, you can operate multi-function calibrators in such a way that they can calibrate 8.5 digit DMMs.
Thanks for the link! That is actually better than I expected, but still not exactly the straightforward operation as described in the manual, and possibly only for some 5720s (since they are relying on non-guaranteed behavior). For example, they report significant errors caused by rearranging an equipment rack. Still, with an advanced characterization procedure it seems possible to use the calibrators for fairly long times (up to a year) with a decent TUR. I stand corrected.
I still don't understand the fascination (obsession?) with calibration for stuff like this.
No argument here. Even if the ranges were out of tolerance by 10x the 1 year spec (something that usually only happens after a repair), it would still be good enough for pretty much all electronics measurements (excluding metrology).
The other thing I'll add here is that the classic Keithley 2015(THD) isn't a nice meter to use for everyday electronics work. The display is very small and the fonts are poor and quite a few of these meters will suffer from a dim/dying display because they will have been run for many hours in an ATE system. The meter is big (as in very deep for a shelf) and heavy and it runs quite warm. The user interface is poor if you want to use any advanced features and this meter is really meant to be used in an ATE system via remote control.
I do not fully agree. Yes the fonts are smaller than some other bench meters (the HP/Agilent/Keysight 34401A and Keithley 199 are both much bigger, for example). And as any VFD, they will get dim if left on for extended periods (people too lazy to turn the display off in ATE applications). But if the VFD was not abused, it is readable in pretty much any lighting and has a much better viewing angle than LCD. So I would still put it above pretty much all handhelds and LCD-powered bench meters (e.g. HP 3478A).
I do not mind the controls. Most frequently used functions are available with a single button press, unlike for example the HP/Agilent/Keysight 34401A which puts current and 4W Ohms as shifted functions. I also like the layout of the ranging buttons better than the Hagisight. Only min/max is quite clumsy compared to how many handhelds implement it (store a fixed number of points, than recall the readings to get the minimum and maximum over that time period).
I wouldn't exactly trust a seller that is part of Micro Precision Calibration [the banner in the background in the picture] selling an expensive piece of test equipment "as-is checked for power up only". A quick google search turned up a branch office in Grass Valley, CA.
I agree that untested is a reason to be suspicious, and could easily describe something that was tested and found defective. But it is possible that the seller did not want to invest the time because the extra money for a 'tested, but uncalibrated' unit may not be worth it. I have definitely bought equipment as 'untested' from seller that should have known better that worked just fine.