Author Topic: BK Precision 4086AWG 80 MHz Arbitrary Function Generator  (Read 980 times)
Offline pienariTopic starter
« on: May 17, 2023, 07:08:34 am »
I have a problem with RF output power.
Have anyone modified full power freq range over 40MHz?.
Full power 20v p-p 0-40MHz over 40MHz power drops to 4v p-p.
I disassembled unit and i found output have an relay array what is cpu controlled.
I tested add voltage to relays but it not raise to output.
This maybe needs mod software?.
Offline Detlev
Re: BK Precision 4086AWG 80 MHz Arbitrary Function Generator
« Reply #1 on: May 17, 2023, 07:23:25 am »
that's no problem, that's according to the specifications:
For all models                 Freq  40MHz: 2mV ~ 20Vpp (open circuit) , 1mV ~ 10Vpp (50)
4084, 4085, 4086          Freq > 40MHz: 2mV ~ 4Vp-p (open circuit), 1mV ~ 2Vpp (50)
This post is "Made in Germany" 😎
« Reply #2 on: May 17, 2023, 09:51:08 am »
Yes i know it works as it should but i need more output power over 40MHz.
« Last Edit: May 17, 2023, 09:54:24 am by pienari »
Online coromonadalix
« Reply #3 on: May 17, 2023, 10:17:58 am »
Wich is normal in a way .. it is designed like this ??
And hacking it wont help,  if the output signal is indeed in specs
You have to add an external amplifier ....
« Reply #4 on: May 17, 2023, 12:07:58 pm »
I think software mod chould be possible solution.
I havent found firmware but i havent ask it from BK.
« Reply #5 on: May 17, 2023, 04:35:14 pm »
Im not aware theses gen series are hackable ....  not like Rigols or Siglent ones who options are hackable, but did not saw anithing related to amplitude levels
and you dont specify  what amplitudes levels you need .. it wont help you get some answers  ...
do yourself some firmware dumps, eeprom dumps inside of it,  and do reverse engineering ??
and with patience  maybe someone can or will help
Offline TurboTom
« Reply #6 on: May 17, 2023, 04:46:30 pm »
Why does this 500ml bottle of beer only contain half a liter? There must be a way to put one liter in, I've got to discuss this with the brewery...  :palm:
Sorry for my sarcasm, but a modern AWG's output stage is quite a tricky design, usually consisting of several, stacked current feedback opamps that, depending on the selected range may or may not get used (thus the array of relays). Offset mixing is often taking place via a separate DAC and passively added to the waveform signal in order not to waste signal resolution, fidelity and to introduce distortion. On these output designs, the factor limiting the output amplitude usually isn't gain-bandwidth product of the amplifier but rather slew rate at the required drive currents.
Though the amplifiers obviously provide some headroom that could possibly be exploited, I'm very much in doubt that you will find any manufacturer of such gear who would support you in an attempt to do so. A dedicated power amplifier after the AWG's output is definitely the way to go.
« Last Edit: May 17, 2023, 05:03:45 pm by TurboTom »
Online mawyatt
« Reply #7 on: May 17, 2023, 04:57:08 pm »
Sorry for my sarcasm, but a modern AWG's output stage is quite a tricky design, usually consisting of several, stacked current feedback opamps that, depending on the selected range may or may not get used (thus the array of relays). Offset mixing is often taking place via a separate DAC and passively added to the waveform signal in order not to waste signal resolution, fidelity and to introduce distortion. On these output designs, the factor limiting the output amplitude usually isn't gain-bandwidth product of the the amplifier but rather slew rate at the required drive currents.
Exactly :-+
Best,
~Wyatt Labs by Mike~
« Reply #8 on: May 17, 2023, 06:25:27 pm »
Well.. i havent found any info from web , maybe BK is so rare.  :blah:
Online mawyatt
« Reply #9 on: May 17, 2023, 07:15:20 pm »
Minicircuits has a large array of various RF/MW amplifiers.
https://www.minicircuits.com/WebStore/Amplifiers.html
You can look around here for AWG Buffer Amplifiers that are a specific for a given voltage, current, power and frequency range.
There's no Swiss Army Knife Amplifier that covers all bases for, frequency, power, voltage, current, linearity, and noise we're aware of, if you find one please enlighten us :-+
Best,
