the newer stand alone LA's are pc's with keyboard + mouse. like the Agilent 1683. hook up external monitor if you want 24". its windows xp.
they also do state analysis + transitional timing and triggering which all these usb things dont do. they just dump X megabytes of data over usb to the pc to analyse.
The 1683 may do some very good stuff the Intronix LogicPort doesn't do (and the LogicPort's memory might or might not be sufficient) but the LogicPort does state analysis, and transitional sampling (and triggering)
- just sayin'
---
http://www.pctestinstruments.com/logicport/faqs.htm
How many full-speed channels does the LogicPort support?
The LogicPort will simultaneously sample all 34 channels with full buffer depth at 500MHz in timing mode, 200MHz in state mode. This is in direct contrast with competing products which sample at full speed on only a fraction of their available channels, or with only a fraction of their marketed buffer depth. When sampling in state-mode, one channel is used as a state-clock input allowing the remaining 33 channels to be sampled at the state-clock rate.
What is the benefit of real-time sample compression?
Real-time sample compression or "transitional sampling" allows the LogicPort to utilize its sample buffer much more efficiently. The data is first sampled, then compressed in real-time using a lossless compression algorithm, then written to the sample buffer. The sampled data remains compressed until the software expands it back to its uncompressed form for display.
This entire process is transparent to the user and has zero impact on measurement accuracy, yet allows the LogicPort to pack much more data into its sample buffer. Compression is particularly effective for high-resolution capture of high-speed bursts separated by long periods of low or slow signal activity. That's because during periods when signals have little activity, the LogicPort can continue to sample at high speeds without filling its sample buffer with thousands of copies of identical data. Only meaningful data is stored in the buffer - memory addresses are not incremented simply to mark time.
The degree of compression varies depending upon the nature of the data, but typically ranges from significant to dramatic. Our proprietary compression algorithm insures that the LogicPort's compression feature never imposes a performance penalty, regardless of signal activity or speed. As a result, the LogicPort enjoys a significant advantage over traditional transitional sampling systems which simply assign half of all available memory locations as storage for time stamps (thereby reducing the effective buffer depth by half for very active signals). To see the LogicPort's sample compression in action take a look at the screenshots, or download the free software and explore the sample projects which are included. As you'll see, it’s not uncommon to collect many times the size of the sample buffer. http://www.pctestinstruments.com/logicport/specifications.htmSampled channels: 34
Timing mode sample rate: 1KHz to 500MHz (uses internal clock)
State mode sample rate: 0 to 200MHz (clock provided by circuit under test)
Sample buffer: 34 x 2048 samples
Maximum sample compression: 2^33 to 1 (sample rates to 200MHz)
Trigger sequencer: 250MHz max, 4ns minimum pulse width
Trigger capability:
Edges - Rising, Falling, Either (multiple channels, any combination)
Patterns - True, False, Entered, Exited (across any / all channels)
Bus Value - Equal, Not Equal, Less Than, Greater Than, In Range, Not in Range
Occurrence Count - Range of 1 to 1,048,576 Consecutive or Cumulative
Duration - Equal, Less Than, Greater Than, Within Range (samples or time)
---
Having said all that, if you can get one of these or a similar model (this is a 1663 rather than a 1683) it looks like a deal:
http://www.ebay.com/itm/HP-Agilent-1663A-Logic-Analyzer-with-accessories-/321884393529?hash=item4af1ce1039%3Ag%3A9JMAAOSw4HVWEaLm&nma=true&si=EMI%252B7m1hl%252FlGFOe5Ntx01RCRabI%253D&orig_cvip=true&rt=nc&_trksid=p2047675.l2557