When I boot up my CRTU-RU into DOS OPSW it complains then:
Please execute FM calibration from the maintenance menu
But in the maintenance menu there are only 3 Options to restore the Windows, DOS or Maintenance Partition...
So where is this FM calibration menu hidden?
Offline fisafisa
Re: CRTU-RU FM calibration
« Reply #1 on: November 06, 2015, 02:41:47 pm »
boot to dos and among the various menus there....
« Reply #2 on: November 06, 2015, 02:45:06 pm »
Well..when I boot to DOS there are no menues then (o;
Offline Wuerstchenhund
  • Able to drop by occasionally only
« Reply #3 on: November 06, 2015, 02:49:23 pm »
"Maintenance" doesn't refer to the recovery partition here.
Boot into the DOS software and go to the main menu, then in the left pane to "Basic Functions" -> "Base" section and you'll see "Maintenance" in the right pane.  This will lead to the Maintenance area where you can do various self-tests and perform a range of alignments (none of them need any tools aside from a RF cable).
« Last Edit: November 06, 2015, 02:53:55 pm by Wuerstchenhund »
« Reply #4 on: November 06, 2015, 03:29:41 pm »
That was it...great, thanks :-)