Author Topic: HP/Agilent 34401A DMM startup problem  (Read 8970 times)
Offline dundeeTopic starter
« on: November 15, 2013, 06:42:25 pm »
Hello there!
I have a problem with my HP 34401A DMM.
It doesn't startup. The display turns on, it beeps and turn off. Sometimes the relais are flipping around.
I already checked the Power supplies, everything seems to be ok. Also i checked the Display Unit - works fine (tested it on another 34401).
Here you can see a Video on youtube, what happens if i turn on the device:
Anybody ideas?
Sorry for my bad English. I'm from Germany.
Best Regards
Steffen
Offline free_electron
    • SiliconValleyGarage
Re: HP/Agilent 34401A DMM startup problem
« Reply #1 on: November 15, 2013, 08:19:10 pm »
Check the ground referenced circuitry in the back around the gpib connector.
It looks like that has trouble starting and it resets the main controller.
These machines ( like the 34970 and many of the power supplies with VFDs ) has a unique way of starting.
the back 8051 initialises rs232 and gpib and runs a selftest. In the mean time the front 8051 (on the display board) does a segment test.
When both are done the main processor starts.
IF either one is defective a restart will occur.
that is what you are seeing.
it looks like something is overheating. after cold start it takes a long time before restart, then restarts occur faster.
check the 7805 regulator in that section if it runs hot.
have you been playing with the RS232 connector ? fried the MC14568 RS232 levleshifter chip perhaps ?
« Reply #2 on: November 15, 2013, 09:14:00 pm »
Hello free_electron!
Thank you for answering.
The 7805 is working fine 5,XXVolt and no overheating.
I bought the device with this fault. Don't know if somebody used the RS232 connector.
Can't find the MC14568, only a LT1180A (Low Power 5V RS232 Dual Driver).
You can hear three "beeps" in the video. This is not always the case. Sometimes I power on, it beeps, display turns on, turns off and that was is.
(no second or third beep)...
« Reply #3 on: November 15, 2013, 09:46:09 pm »
are the main firmware roms in a socket ?
pop em out and reinsert them.
resolder the socket pins on the board if needed.
it is your main cou not starting within the allowed time and the i/o cpu pulls the reset line
« Reply #4 on: November 15, 2013, 09:55:46 pm »
No socket's :(
« Reply #5 on: November 15, 2013, 10:50:18 pm »
whoa. what is that big blister on the circuit board ?
between the big intel processor and the white 3 pin connector, just above the little 8 pin fairchild chip.
it looks like a round waffle . that is board delamination !
check those interconnections in that area. you may have a broken via in there around the capacitor or the pins of the big cpu.
especially since a trace goes to the optocoupler that carries signals to the 8751 in the back portion.
if that fails you will see those whacky resets...
« Reply #6 on: November 15, 2013, 11:03:47 pm »
That's not the mistake... I resoldered the CPU with a heatgun, then the blister apperars.
The failure was the same before the blister.
« Reply #7 on: November 16, 2013, 01:03:23 am »
 :palm:
so you fried the board now too ...
Offline razberik
  • Country: cz
« Reply #8 on: March 06, 2017, 01:48:15 pm »
Hello. Sorry for zombie-thread-digging, but I wonder - did you resolve the problem with the DMM ?
Thanks.
Offline SPAZIONICA
    • www.spazionica.org
« Reply #9 on: December 25, 2017, 09:51:58 am »
Hello everyone,
my HP34401 had the same problem,
I solved by re-doing the welds in the back area PCB in particular I think those of the quartz
see attached jpg
Bye Bye
Francesco
« Last Edit: December 25, 2017, 09:57:36 am by SPAZIONICA »
The following users thanked this post: Mr. Scram