I just tried adjusting the autoprobe resistor on two 1152A probes from 56.2k to 14.7k (for x10, 50 ohm) by putting a 20k in parallel with the existing 56.2k, and it seems to work nicely in terms of attenuation accuracy (I measured around 0.5% DC accuracy with a 34401A).
The offset seems to depend on the scope you're using not just the probe itself. For the vast majority of probing I do, the offset is fine on a 54831D (that detects an 1152A out of the box) and MSO7104B. I wouldn't bother using it on an MSOX3054A, a 400mV offset is going to confuse things.
Offsets:
Scope Probe 1 Probe 2
54831D -17mV -31mV
MSO7104B +29mV +12mV
MSOX3054A +370mV +410mV
The peak DC voltages before saturation or, for negative voltages, overload protection kicks in, are below:
Min/MaxV:
Scope Min Max
54831D -13.5V +12.5V
MSO7104B -10.5V +15.7V
Open up the 'scope end of the probe lead by
gently prising the four visible plastic clip points one at a time, while separating the upper and lower parts of the enclosure as appropriate. It takes a little while to get the knack, and those clips are asking to be broken (if they're not already).
Then put an 0805 20k 1% resistor on top of the existing 56.2k resistor... see pictures.
Before...
After...