I wanted to bring to attention what I believe is an example of drifting or failing calibration. In a specific sense, none of this matters (because you are probably going to
normalize, which makes the measurement you make relative to the problematic measurement below). It's still interesting, and I'd love to hear speculation on why this is happening. I can make further measurements (such as using a different spectrum analyzer to view the output, or a SDR).
The following will tell you the default settings after preset, then show which one or two things to change from the previous. For all tests, the only connection is a 0.5 meter Agilent N-to-N cable.
Two Port Insertion Loss (default settings)
* 375 MHz - 2.5 GHz
* Source Lev
Auto (10 dBm)* Optimize
AccySo when we go to a preset configuration, we have a sudden downward jump at 1 GHz. In addition, the average insertion loss is genuinely lower than it should be. Substituting a different cable has no effect:
If we pick a manual output level, then we can cut down on this effect...
* 375 MHz - 2.5 GHz
* Source Lev
Manual (0 dBm)* Optimize Accy
And if we keep reducing the output level, we can eventually make this jump very small:
* 375 MHz - 2.5 GHz
* Source Lev Manual
(-10 dBm)* Optimize Accy
If we change to "Range" mode, this sharp jump effect totally disappears but the insertion loss offset
does not change.
* 375 MHz - 2.5 GHz
* Source Lev
Auto (10 dBm)* Optimize
Range
Likewise, if we change to Manual source level and change it to - 10dBm, we get a more or less correct measurement of insertion loss:
* 375 MHz - 2.5 GHz
* Source Lev Manual [/b](-10 dBm)[/b]
* Optimize
Range
In case you think this is an effect which shows up only when you have a large span, it's exactly the same up close:
*
999 MHz - 1001 MHz* Source Lev
Auto (10 dBm)* Optimize
Range* Source Lev
Any thoughts, guesses, or further tests would be appreciated. I have SDRs which I can use to monitor the output from the CW/generator. I can use another spectrum analyzer to view the output. I think I've done this, and I didn't see any sudden jumps at 1GHz exactly. Likewise, when I use another signal generator into this spectrum analyzer (E7495a), I similarly do not see any sudden jump. So this leads me to believe that this effect is from some sort of calibration table that the unit stores for the Two Port Insertion measurement mode specifically.
It's worth noting, too, that the procedure for this originally calls for using a two 10dB attenuators (back to back) on the RF out port. Using an attenuator set does, of course, reduce this effect but it's still noticeable.