Hello Volt Nuts,
Update of the HP3458A story.
I would like to thank Quarks for he’s kind support
and all others
I could manage finally to get the instrument.
It is in a fair cosmetic condition, only small usage marks, after removing all labels and glue from calibration labels, the front panel is in mint condition. I hate all those labels on the front panel and sometimes when the instrument is defect they put it on the display.
The LTZ1000 is covered by a plastic cap, is this done on all HP3458A because I could find some pictures on the net with the option 002 with this kind of protection?
I order 1 DS1220 and 2 DS1235Y, I will change them to DS1230Y.
I order a Geller Lab SVR-T that is on he’s way to spend one week in our customs.
I order a 10K resistor 0,005% / 0,05ppm/°C
I removed the 3 NVRAM without problem with my iron with pump, easy job, just 2-3 minutes for 1 NVRAM, no damage on PCB *.
Put 3 industrial gold plated sockets and solder them on the PCB and clean it.
I checked the DS1220Y by reading with my eprom programmer and is virtual empty, just a few random values. By using the test function it give an OK
I don’t really know what they do in the testing procedure, I suppose reading content, saving it to put back after test, writing FF, reading again and comparing if it is the same and write back the saved data.
By testing the 2 DS1235Y my programmer report that one is bad at a specific address.
I got the ordered DS1220Y from Germany last week and put it in place
, still waiting for the DS1235Y from UK.
I try to put the DS1235Y(s) back and depending of the position it give me an error HIGH or LOW and it could be logic if one is bad and detected by the test procedure.
I upgraded the firmware as well but nothing changed, instrument is still only accepting the SHIFT and RESET function.
I hope that there is no other issue after adding the 2 NVRAM.
I intend to do the following to calibrate it:
1/ I have 2 calibrated in metrology lab 6,5 digits meter, Agilent and Picostest (They have similar specifications), I have a DER DEE for the resistor test.
2/ I will measure the value (after warm up) of the resistor with the 2 6,5 digits meter and the DER DEE and make an average and this value with the resistor will be used to calibrate the HP3458A.
3/ Start the calibration with the copper wire, the 10V reference from Geller lab and the 10K precision resistor.
4/ I hope that this should finalize it. I will check the Agilent notes about the instrument.
I will update when I get the 2 missing NVRAM.
* Used wrist strap and safe table connection to ground.