Hi to all, i recentrly obtained two multimeters in bad condition, with different board revision, one with v05 and one with v11. I Have repaired all of these, with this occasion i have analyzed the differences between two models and i have extracted also the two firmware of the comm cpu v1 and v5. In the beckdating section of the service manual i have verified the difference and the result is this:
first change board revision
C111 100 pF ±5% 50 V add
C402 220 pF ±5% 50 V add
R327 21.5K ±1% .125W add
R403 8.25K ±1% .125W replace 12.1k
R422 46.4 ±1% .125W add
R430 Zero Ohm replace 14.7ohm
U102 custom resistor, i think is only a different production, i have measured the two different codes without desoldering and is very closest
second change only components
CR203 Zener 5.1V 5% replace 4.7v
R207 2.37K ±1% .125W replace 5.62k in manual, but mine is 1.78k, other in photos are 1k
other changes is only production variation, the last important change is the dipslay panel from version 07 and the firmwares of main and comm cpu
i have analyzed the differences and i have founded this:
The gate of mosfet q301 is connecet thrugh the r327 resistor, i think it work only in saturation and the resistor decraese the stress on the switch ic.
The power to U411 and relative cap C410 is delivered by u401 5v reference buffer thrugh the r422 resistor, instead from u553
Replaced the zener voltage that power the ohm current generator from 4.7v to 5.1v
The pin vee of the mux U411 is connected to gnd only with the inductor, insted the r430 in series
Other changes in the amplifiers chain of the final stages that go in the adc.
Someone think that all this changes can be do on the old pcb for update all hardware without concequences? Improve something? Obvius that a recalibration is necessary.
Here is the firmware of comm cpu u700 that i have dumped
http://www.ko4bb.com/getsimple/index.php?id=manuals&dir=HP_Agilent/HP_34401A_6.5_Digit_MultimeterIn the latest revision of the pcb is present a jumper that connect to ground the a4 address line of the ram, if the jumper is removed the gnd disappear from signal and remain connected to the cpu pin 27 that is an high speed output, on the old revision this pin is grounded and the a4 line of the ram is connected to vcc, the result is the same, a part of the ram is unused. Is interesting if on the firmware effectively is supported.
regards
gerry