I just got a Solartron 7060 from an auction in an unknown state. First success is it was a full blown G model with V/I/R features. Then I turned it on and turns out only the resistance wasn't working properly. Winner!
After a solid 4 hours of troubleshooting I managed to narrow the resistance reading issue it down to a single reed relay that wasn't switching the current source to the I- bus. Swapped that out and it was happy. Until I went to plug the board in and intermittently it would show the resistance and current modes as "no function", aka it hadn't recognised the board that handles those functions was plugged in. Narrowed this down to the old school pin connectors Solartron seemed to love in that era and made a shim to go under the PCB to hold the pins in place. The connections were now totally fine. Unfortunately during the process or putting it all back together something went awry and I'm now looking at a screen like the following:
Normal boot:
Then it sits on the two dashes. Not taking any measurements.
The sample LED on the analogue PCB doesn't tick at all making me think something isn't triggering any samples to be made.
At this point, a full day of debugging and getting so close to a working device, I'm taking the evening off and having a beer. I'm posting this on the off chance of some sympathy or that someone has seen this failure mode before. Cheers all.