Author Topic: SDM30XX, Input terminal protection limitation  (Read 831 times)
Offline knudchTopic starter
« on: November 04, 2021, 07:01:00 am »
In the Quick Guide for the SDM3065X on page 6 is written:
"2. Sampling (HISense and LOSense) terminal
HISense and LOSense are used for 4-wire Resistance
measurement. Two protection limitations are defined:
 HISense-LOSense protection limitation: 2000Vpk.
 LOSense-LOSense protection limitation: 2Vpk."
Specialy the line LOSense-LOSense.......
What does that mean ?
Offline Kean
    • Kean Electronics
Re: SDM30XX, Input terminal protection limitation
« Reply #1 on: November 04, 2021, 07:53:38 am »
I would say that is a mistake and it should be LOSense-LO is limited to 2Vpk
« Reply #2 on: November 04, 2021, 08:15:03 am »
That was also my first thought
But that means it is only allowed to have 2Vpk between LOSense and LO ?
In case you cross HISense and LOSense (Mistake) and more than 2Vpk HISense-LOSense will it then destroy the meter ?
« Reply #3 on: November 04, 2021, 08:20:07 am »
Keysight 34460 has 200Vp limitation between HISense and LOSense to any combination
« Reply #4 on: November 04, 2021, 08:28:38 am »
Rigol DM3068 has 0,5Vp limitation LOSense - LO !
« Reply #5 on: February 24, 2022, 10:47:51 am »
Finally I got an answeer from Siglent Europe
Quote:
"I got the final feedback that our value should be like at the Keysight you are referring.
The manual will be fixed soon."
I understand it like this:
"Keysight 34460 has 200Vp limitation between HISense and LOSense to any combination"
We will see when the manual is updated