I tested a uCurrent Gold with a 400Hz AC signal, to get to grips with the lowest current I can measure with it. I used an AC measurement to be able to filter out noise.
Setup: Signal generator output 50mV RMS AC at 400Hz driving a precision isolation transformer 1:4 ratio, in turn driving a 7 digit Ratio Transformer at 200mV RMS. Star ground point at the uCurrent negative output terminal to reduce noise.
uCurrent set to highest sensitivity, 1mV / nA. Measurement taken with Agilent 54622D oscilloscope set to 512 averages, to reduce noise.
Results -
Lowest visible current with decent measurement accuracy (3%):
600pA at 400Hz
Calculating the current: Ratio Transformer Drive Signal (200mV RMS) * Transformer Ratio (Dial) / Shunt Resistance of uCurrent
0.2VRMS * 3.0E-5 / 10Kohm = 600pA
Lowest visible current - but with poor measurement accuracy:
60pA at 400Hz