well there been one week since I began to work for the cal lab . I am astound from the IQ floating there ( the customers , not employees)
they send me little palm sized Uni-T for calibration , WTF ? , that Uni-t cost few dolars , the calibration cost 3X more the meter himself !!!
Just today I've checked some 1.5KW Lambda supplies , they all didn't pass , one of them gave me -7V while the set voltage was 2V ! ( always turn your kikusui E-load off when enabling the PS output ! , you might stumble around supply who gives you negative voltage
, and then poof your E-load 0_0 )
Dave - theres nothing to the country , the thing who gets me crazy is the stupide educational system , there are "EE" degrees everywhere ! , but they are so dumb
today I actually hosted some guy how to use the scope
he freaking marked on the kobes the static position for the particular use (analogue scope of course ), he has degree but can't use scope ? 0_0 WTFFF ?
BTW , today I calibrated HP3458 , with the fluke 5700 , I set the thing to 10V . the 3458 says " 10.0000142V - that's 1.42ppm out , there was also keithley 2001 , I set it to 7.5Digt and he also showed 10.000014V ( aint it amazing ? ) so I told my boss that he ought to che?k that out , he must calibrate the fluke 5700 ASAP . NIST traceable of course ,
he might sell me that keithley for around 250$ ( yes I didn't forgotten one zero or so
)
he think thats old meter and so on ><"
who cares ! that's good meter ! , it'll be my reference for my lab ! ( I'll use it as 6.5Digit absolute value )
I really coursed out of topic there , sorry
I guess I wanted to say that you perhaps right , I know too many abusers over here , and very small community who really know what's T&M's value