So I picked up a few 5316A's recently, all of them with the 1 GHz C-channel input(Option 003) - one has what is called the C10 option. It is not mentioned in the manuals I have found, nor in the old catalogs. It is mentioned in two places in all the internet that I have found so far, and is described as a parallel connected 50 ohm input for both channels A and B, and that they should be terminated when not in use. While the C channel seems to do fine in testing down to about 1 MHz, it is listed as 50-1000 MHz, so I suppose the 50 ohm input for the A channel will be handy since I mostly do radio work.
I just find it odd that I can't find any official documentation or other mentions of this option anywhere. Does anyone recognize this option or have experience with it?
I have attached some photos, showing the rear input BNC's(seems a little close to the AC power input), the route of the coax through the middle, and where it is soldered to the front panel inputs.
The only issue I've run into is that two of these units showed up with errors - one I'm still working on, the other started working after reseating all the socketed IC's and fiddling with the external reference switch.
ETA - Third unit fixed. Note to self - don't use electrician DMM that only reads to tenths of a volt. 5V rail was actually 5.14v, spec says 5v +/- 0.01. Once adjusted using my o'scope, the counter was perfectly happy. I really need to hurry up and buy a good bench meter. So if you have one of these with the GPIB LED's flashing in an alternating pattern and the external reference switch isn't dirty or stuck, check your voltages carefully!