yes i did it of course (yello and blue while turning on) but does not end , shows "Err0" after some seconds displaying those weird numbers...the pots are back at at same place but i think still differs from the BIN file i have downloaded... so wanted to match them somehow...
I have also found this procedure below very interesting!, can anyone tell me if this guy used a known value resistor connected to do the calibration process:
Calibration must be done with caution. I am not responsible for any problems. Remember, remember, remember to put
the turntable in the OFF position, press the blue (SELECT) key and the yellow (HOLD) key at the same time, and quickly rotate the turntable. Turn on the clamp meter to the Ω position until the CAL character is displayed on the screen. Release the blue and yellow keys. At this time, it has entered the calibration mode
1. After that, the machine will automatically check the internal circuit of the IC. The LCD will display the ADC code value. If there is an error, it will prompt Err0. ~4. At this time, turn off the machine and check whether the voltage dividing resistor for measurement is connected to see if there is a short circuit or open circuit, and whether the resistance value is correct. After the check is complete and there are no errors, restart the machine to start the calibration process. After the self-test is completed, the relevant parameters will be automatically saved to E2, the buzzer BEEP will sound once to indicate completion. If it has already been tested, press the SELECT key to skip this check.
2. Then automatically check the relevant parameters of the internal amplifier. After the self-test is completed, the relevant parameters will be automatically saved to E2, the buzzer BEEP beeps to indicate completion. If it has already been tested, press the SELECT key to skip this check.
After pressing the select key twice to skip the above two self-test steps, the turntable must be rotated to other gears before the calibration is official. Start (this is very important)
3. Turn the turntable to the gear that needs to be calibrated, and perform calibration (it is recommended to calibrate the DC mV gear first)...
During the calibration process, the function of pressing the blue (SELECT) key is " "Skip this step", short press the yellow (HOLD) key to subtract (-) the calibration value, long press the yellow (HOLD) key to add (+) the calibration value; if you want to exit the calibration mode, turn the dial to the OFF position."