Anyone willing to test two new eeprom mods? It gives two new functions in 2A mode. Its reported on the Russian site that these functions still have the full count even after zeroing out (i.e. doesn't go back to 2000). The LCD does not show a "." in the read out.
Update: Tested the DCA 0000 function in 2A and it truly works even after zeroing out.
old new
A7: 00 1C - DCA 0000 in 2A mode / SELECT slot 3
B7: 00 1D - ACA 0000 in 2A mode / SELECT slot 4
That group is really digging into it. So is the decimal gone as they suggested?
With my programmer working again, I tried these changes. Currently I have the meter default to DCA/DCV.
If I turn on the 210E, then select the 2A mode, the decimal is shown. Zero out the meter, the decimal is still shown. Putting more than 2A through the meter will over range it.
Turn the meter back off and on, then again select the 2A mode. Don't zero out the meter and place more than 2A through the meter. It does not over range and reads the correct value.
Turn the meter back off and on, then again select the 2A mode. Select AC, the decimal is shown. Now select DC, the decimal goes away. Don't zero out the meter and place more than 2A through the meter. It does not over range and reads the correct value. However the meter now starts beeping at about 600mA.
Turn the meter back off and on, then again select the 2A mode. Select AC and then DC. Zero out the meter and place more than 2A through the meter. It does not over range and reads the correct value. However the meter still beeps at about 600mA and there is no decimal.
So it's not a real clean setup IMO. Maybe I missed something.