I just received my scope last night and spent a couple hours getting familiar with it. So far it's great!
I recall watching a video by one of the popular youtubers (mjlorton, I think) where he was using a scope to measure the voltage of LiPo battery while being charged. This was a long term measurement (hours, I think) and I wanted to make a similar measurement with my DS1054Z. Can it do it?
This is where I got confused about the waveform record, which I thought I could use. Here is what I thought given the numbers in the manual:
Interval max is 10 s
Frames (or segments) I thought I read somewhere was 5000 max (but this is not the case as it appears to be a calculated max)
Using my first assumption, I calculated 10 s * 5000 = 13 hours. I thought I would set a 10 s sweep, but then record is not available.
If sampling rate is set to 12 kSa/s, then with 1 GSa/s, you'd get 23 hours, so I thought I'd use the Roll mode - but again no recording is available (which I then found the recording limitations in the manual).
Now I am also aware of this formula:
MDepth = SRate * TScale * HDivs
and that is indicated for each Frame (which they actually call Length).
This seems to be a limit:
Mdepth = 24,000,000
Srate = 12,000 Sa/s
Hdiv = 12 divs (a constant)
Tscale = 0.000165 s or less
per frame.
I'll have to work out the math on some of my settings, but I came across a few scenarios where I could not increase the Interval to more than or less than 100ns and only 2 Frames, and my time base was set around 8 MHz to probably around 100 kHz (I was looking at an 8 MHz 5-pulse burst with a repetition rate of around 1 MHz). I do not have the scope with me now, but these are two things that did not get completely resolved on my first run-through of knob twisting...