https://www.eevblog.com/forum/repair/tds754d-not-booting-anymore/Well that didn't last long...
Did some test worked great power off the scope, take a 1 hour eating break. Baby sleeping still have time to do more test, power the scope go to the bathroom 2 min come back, all the led still on, no screen, no response. Reboot same thing....
Reopen the scope led display stuck to .8, if i cycle the unit fast sometime i have a .E , or a .8 .E .8 sequence showing very fast or a weird c (a,f,g segment on).
what i did before (for anyone not following this post), Hacked the scope for the 2m 2c option, worked great for a month till today, replaced the ds1250 Nvram by a fresh one with the backup of the old one.
Troubleshooting.
2 fan blowing on the scope case open.
-Tried to cycle the scope many time. Same thing
-Replaced the fresh nvram with the old working one. Same thing
-All the cable are tight in the right slot.
-Waited 5 minute, same state, tried to talk via GPIB, got an error, the scope dosen't reply..
Start to read the service PDF.
-All voltage OK
-All voltage look good on my other scope.
-Frequency good at the pin 25 of the J2 connector.
-Tried the s1001 switch with the help of some other post on the Tek forum. Same thing.
-Tried many combination with other board removed. Same thing.
With all the troubleshooting and following the flow chart the only thing this manual screaming is Replace the A11 DRAM Processor/Display module.
Checked the thermal signature with my E4 and nothing weird show up... No cap leak and nothing out of place visually.
I have the impression that the cpu is stuck on reset or briked. I tried to probe the Nvram to see if the CPU toggle the CSn pin and nothing happen.....
damn.. never again ... if it ain’t broke, don’t fix it
For now i have a 16 Kg paper weight, if someone have an idea....