I’ll follow this up with a more detailed review and possible teardown of a BK 5491B DMM (50,000 count).
I want to make sure that I understand the specified accuracy of this meter on DC voltage.
This specification is slightly different from Dave’s video:
5491B DC Accuracy = +/- ( 0.02% of reading + 0.008% of range)
So, if I measure a 10.000 volt source, using the 50 volt range, the reading should be:
+/- (10.000 x 0.0002 + 50.000 x 0.00008) = 0.002 + 0.004 = +/- 0.006 volts
When measuring a 10.000 volt source, the reading should be between 9.994 and 10.006, right?
What I’m getting is a very, very repeatable: 10.018 volts. When I reverse the leads, I get -9.982 volts. This seems way out of spec, I originally questioned the accuracy of the voltage source, but not any more (I’ll post more on that later). When I short the leads, the meter reads 0.000, could the internal voltage reference be way off?
I appreciate any thoughts on this.