Note that while both my DMMs are on 2.09, they actually show slightly different firmware versions on the About screen:
34461A: A.02.09-02.37-02.09-00.49-01-01
34465A: A.02.09-02.37-02.09-00.49-02-01
Not sure if it is a firmware difference. I am not sure if I updated both with the same file.
No the firmware is the same, that number represents the measurement board revision - the version number is explained in the section about the *idn? command (page 270 of the current (rev 3) manual). Just for completeness, the version format is
h.ff.ff-pp.pp-mm.mm-gg.gg-bb-fp
where,
h - Hardware revision
ff.ff - Main firmware revision
pp.pp - Front panel firmware revision
mm.mm - Measurement board firmware revision
gg.gg - Measurement board FPGA revision
bb - Measurement board revision
fp - Front panel board revision
and it is your bb value that is different
I suspect that the difference is due to the measurement board changing, with the newer firmware using a lower threshold value, which doesn't work as well for the older revision boards. You can guarantee that they develop the firmware on the highest model (i.e. the 34470A) and just change the model number to cap the features - I doubt they have got all their firmware engineers with 4 multimeters stacked on their desk.