Hi all,
Recently, a viewer of my YouTube channel asked wonderer if the GW-INSTEK GMP-8310 could be used to measure inrush current (a.k.a. switch–on surge, or surge current). This instrument does not have a specific measurement mode for this (and the manual does not mention inrush current measurements), but it still may be possible, I thought. After all, this device has sufficiently fast DACs (after all, for a signal with a fundamental frequency of 67Hz it can determine up to the 50th harmonics…), so, in principle, it should be able to pick up fairly short spikes. I was also interested in this, also to see whether we could determine, for a given DUT, how inrush current depended on the power supply phase point when powering on. But anyway, let’s start with the basis.
Today, I did a couple of experiments and it seems to measure inrush current is not that easy indeed. As a DUT, I used a USB-C power adapted that shows visible spikes when I put it into an outlet, so I expected to see some quite notable inrush values.
Firstly, the device does boost an ‘MAX Hold’ mode. When playing around with this, I realized that this mode only takes the max values at the maximum screen update rate of 0.1s. Not sure how the values in this interval are exactly determined, but it seems its not reporting the highest peak in that period.
Second, I tried the data logging mode, as described
here. Also this way, the values we get are limited to a maximum screen update rate of 0.1s, so again, not showing the real peaks.
Third, I tried the Windows software utility. It's quite a challenge to get the interface working (I never got the USB working; did eventually get the LAN connection going after many error messages). With this utility, logging gets a bit more interesting. It saves a set of 8 different files:
- One file contains measurement values at (again) measurement intervals of 0.1s or higher, and an associated graphics file is added with the voltage component.
- Another file contains measurement values of the 50 harmonics, and two associated graphics files are added with bar graphs for voltage and current harmonics, respectively.
- And finally, one file contains higher sample rate measurements of voltage and current, and two associated graphics files are added with graphs for voltage and current (oscilloscope-like waveforms). For these measurements, I observe about 240 measurement values for a single cycle of a 50Hz signal, so about 12k/s. Great, probably good enough to capture inrush current. But there is a problem. We get only 600 measurement values in total (so about 2.5 cycles). As far as I can see, these represent the last couple of milliseconds of the total measurement interval. If you wish to observe an inrush current here, you somehow would need to figure out how to turn on the DUT in the last few milliseconds of the measurement interval, an interval which you need to set in advance in the Windows app.
So, all in all, I did not find a doable way to measure inrush current yet with the GMP-8310 yet. Too bad, as its hardware should be able to do this, and this type of measurement is available in some other digital power meters. Within the same price category, the R&S HMC8015
has such a mode, and also the Itech IT9121E can measure inrush currents, according its specifications
Maybe I’m overlooking something and someone knows how to do this? If so, let me know!
Best, Rudi