Looks like a hardware issue.
I drew the schematic diagram of this clone because I have this device and have studied all its features well.
In red color I showed the improvements that are very necessary to make so that the device works stably!!!
It is especially important to replace the weak and low-quality stabilizer HT7150 with a stronger MCP1702!!!
C10 can be removed from the circuit altogether, its absence will not affect the operation of the circuit, but C12 is important and preferably of high quality!!!
Yuriy_K made a correct comment on the calibration procedure. Initially, you need to disconnect all long conductors and calibrate with a short jumper and then check the details on the same contacts. If longer conductors are needed, then the calibration procedure must be repeated and the parts checked only on these contacts.
It is also very important to correctly complete all calibration steps and use high-quality calibration capacitors, because the correct operation of the device depends on this. All these nuances are discussed in detail in the author's manual, read very carefully, otherwise you will constantly have questions and problems.