Hi, Madires
I am classifying SMD components and have noticed a failure in your firmware.
Observed in several components, a problem that does not occur with the K firmware, nor with the Chinese firmware that incorporates the
GM328 or LCR-T4 meters.
An example in photos with a transistor C5707 for the test, in the datasheet says that it is a NPN Bipolar transistor.
The original Chinese GM328 and LCR-T4 meters, and the GM328 with the K firmware, the transistor put it in position
that you put it tells you that it is a bipolar NPN transistor, and it appears as the datasheet says, everything is fine here.
Now the same transistor with the GM328 and with the firmware M, in one position it appears as a NPN Bipolar transistor, and in the other position it appears as a JFET, leading to error by not knowing positively what the component really is.
Look at the photos and look at the position of the transistor that is only rotated in the meter socket.
The same applies to the double SOT23 three-pin encapsulated diodes, connected to the original Chinese GM328 and LCR-T4 or to the GM328 with the K firmware and appears as two diodes, either in series, opposed or as they are inside the component but always
It detects it as two diodes.
With the firmware of yours, you connect the three pins and it appears as a single diode or as a JFET transistor and it costs you
much measurement and in some cases it gives error.
I tell you because I do not know if it is my problem and if it happens with other components, and if not, if this problem could be solved and corrected.
Thank you very much.
Cheers