I repeated all of my measurements and recalculated the coefficients with METAS. As before, poor performance.
....
Still, even feeding the METAS results back into the PNA, then calibrating with the low cost standards, results are poor. I think the writing is on the wall, we need better standards.
I had gone back over all of my work a second time and noticed when I had used the calibration editor to create the file for the Agilent. I had transposed the 2nd and 3rd coefficients for both the short and open.
After correcting the file, I calibrated the system again using the V2Plus4 standards with sorted load. This time when I measured the T-Check, it was well within +/-10%.
Also shown comparing the S-parameters of the T-Check when measured with the E-Cal and with the METAS fitted model.
It certainly suggests even these low cost standards are not my limitation with the LiteVNA. Yes, I rechecked to make sure I hadn't transposed the data in Solver...
***
Note, the error can be seen in the previous post. The older CalEditor uses the same arrangement. I think being so used to reading left to right, top down, I could look right at the problem and not see it. Too funny.
So far, this was all done using Agilent's firmware/software to avoid adding more complications by introducing my software into the mix. Next step, update my software for the Agilent and see if I can get the T-Check to perform well.