With much thanks to Dislord, I now have a working version of firmware. I don't think any of us are expecting anything from the LiteVNA at these higher frequencies and I think we all agree that limiting it was a good choice. However, I was a bit surprised and it performs better than I was expecting.
8G_12G_Leakage_SO_10dB:
Showing a thru (yellow) and open (violet) after calibration. I want to be clear, this is still the ideal model using the same standards provided with the V2Plus4 with the sorted load. Also, this is my unmodified LiteVNA right from the factory but with my cables. My software accounts for the leakage term and it is being used for these tests. A 10dB attenuator was then measured. We can see is starts to go off the tracks a bit over 9GHz but settles back down around 10.5GHz.
8G_12G_Leakage_SO_20dB:
Same as above but with a 20dB attenuator.
I have been experimenting with some waveguides using the LiteVNA. Most of this testing has been above 7GHz. To work around the limitations of the LiteVNA, I use it below 4GHz and then up convert to 8GHz then back down. This setup allows me to take advantage of the Lite's higher performance range. test14_MWM_0_50dB_11GHz shows the Lite with the Extender sweeping from 11.0 to 11.9GHz and inserting a series of attenuators from 0 down to 50dB.
8G_12G_Leakage_10GHzBP_WR90:
Showing the Lite directly measuring a 10GHz waveguide BP filter I put together. The last plot was measuring this same filter with the LiteVNA plus the frequency extender. I don't have the equipment needed to test this filter. The PNA is limited to 9GHz or I would use it for a comparison.
I'm impressed, we can actually see something! Pretty cool for something that cost $120 and fits in my shirt pocket.