I purchased recently one HP478A thermistor sensor without the calibration chart. The sensor is working properly and when I compare the results at 50 MHz with my Rigol DSA815 spectrum analyser, the measurements are close together. However, the question is if the results will be correct on the higher frequencies. Can someone kindly help me with suggestions/ideas on how to resolve this?
Probably the first thing to do would be to measure the return loss of the sensor across the frequency range you intend to use it over. This should provide a fairly good health check of the sensor.
At 50MHz it should typically be better than 25dB, but by 100MHz it should typically be better than 30dB and I'd expect it to maintain at least 30dB return loss to at least 1GHz.
However, it will degrade above about 1.5GHz. I don't recommend using the 478A above about 2GHz because the return loss isn't very good here. It is possible to fit an attenuator to the sensor to improve the return loss above 1500MHz, but you then need to know how accurate the attenuator is.
Obviously, it is also risky to use this power meter below about 30MHz because the return loss degrades fairly rapidly below about 50MHz. You can improve this with a (say) 10dB attenuator fitted to the sensor, but you need to know how accurate the attenuator is.
If the return loss is as good as I've listed above, then the sensor is almost certainly in good condition and the cal factor will typically be 99% +/- 1% across 50MHz to about 1500MHz.
For most home users, I'd recommend doing the return loss check on a decent VNA and then simply leave the cal factor set at 98% or 99%.
This ought to be good enough for all but the most critical applications. If the critical application is above 1500MHz, then I'd generally recommend using a modern thermocouple based power meter, unless the source impedance of the DUT has a very low VSWR.
I'd expect the cost of calibration to exceed the cost of buying another used and tested sensor, although the prices of these sensors seems to be climbing every year. Prices also seem to vary from £150 to over £1000 for a tested sensor.