Author Topic: CMU200 repair attempt - boot error, stuck on "stuc.cpp".  (Read 5199 times)
Offline YansiTopic starter
  • STM32, STM8, AVR, 8051
Re: CMU200 repair attempt - boot error, stuck on "stuc.cpp".
« Reply #25 on: October 28, 2020, 08:18:51 pm »
Hello,
I am currently still not at home, where I have my lab and all stuff. I'll be back likely on friday.  Need to disassemble the module back again and desolder the 16.2 ohm resistor. Only then we can distinguish what the values of the two shunt resistors are. Will report back then as I've promised.
Y.
The following users thanked this post: YU7Q
Offline YU7Q
  • Country: cs
« Reply #26 on: October 29, 2020, 05:16:28 pm »
We look forward to your reply.  :-+
Offline Ice-Tea
    • Freelance Hardware Engineer
« Reply #27 on: October 29, 2020, 06:25:18 pm »
Two pointers for backlight replacement:
- as far as I remember, there's only 5V on the BL connection and 12V is not readily available anywhere near. Most LED"inverters" I've seen are 9V+
- The CCFL itself can be found as well, but it's an L-shaped tube so no entirely straight forward either...
« Reply #28 on: October 29, 2020, 06:53:21 pm »
I plan to design a custom switchmode constant current boost converter for the LED backlight, that would fit in the original place where the CFL inverter lived.
I've just arrived home, so tomorrow I'll start working on it.  Need to identify the original BL connector first.
Offline YU7C
  • Country: cs
« Reply #29 on: November 15, 2020, 08:02:52 am »
About RF4IN broken input attenuator:
I put only 18.2 Ohm serial resistor (no shunt resistors), and got T-attenuator, with a good accuracy (within 0.5dB) up to 200-300MHz, and about 2.0-2.5dB lower results on 1GHz.
Also measured RL is better than 30dB on freq. up to 500MHz, and about 25dB on 1GHz.
Still no information about shunt components.
Vojislav
« Reply #30 on: November 22, 2020, 04:13:23 pm »
Ah, now I see I did not post my findings  :palm: :palm: :palm:
The two shunt components, or at least the one near the input, measures >20Mohm, after desoldering the 16.2ohm resistor.  Measuring across the other shunt component (with the rest of the instrument following it), measured 82 ohm, almost spot on.
I have absolutely no idea, what that high resistance component is. But it seems is not a resistor.
Same component used in different places, even near the P3 port.
If you look closely, it even does not look like a regular resistor.
The following users thanked this post: YU7C, YU7Q