Author Topic: Cambridge Instruments – Dolezalek Quadrant Electrometer  (Read 19000 times)
Offline GreggTopic starter
« on: September 02, 2018, 08:42:12 pm »
What an amazing source of information the EEVBlog is; this was resolved in a couple of hours.  This post was originally posted asking “what is it”.  Forum member chris_leyson came up with the answer.
It is marked:
   Cambridge Instrument Co. Ltd. England
      C 157555
      Hand engraved: IR9450
The overall height is about 260mm. Diameter of the removable cylinder with the visual port is 95mm; it is painted flat black inside.
At the top in the center is a pin that moves up and down with a sharp point on the bottom.  It does not have a place to make an easy electrical connection like the other three protrusions on top.  It is electrically insulated and used for hanging a butterfly shaped disc that fits in the hollow of the quadrants below. The stem that is missing should have a small surface silvered mirror that will reflect a light beam through the window in the cylinder.
The other three top terminals are all insulated from each other and the frame; they have holes for wires with thumb screws to fasten them.
The little adjustable whisker is to add or remove a small charge to the hanging butterfly disc.
« Last Edit: September 05, 2018, 04:07:12 am by Gregg »
Re: Cambridge Instruments – Discharge or HF Chamber, another WHAT IS IT?
« Reply #1 on: September 02, 2018, 08:44:34 pm »
More pictures of the Cambridge instrument:
Some provenience:
I became the executor for the estate of a friend that worked as an instrument maker who made lots of neat things for the physics department of a major university.  My friend also collected things that he thought were neat; I’m sure most of us on this forum can relate.  This instrument was probably something that he thought was worth saving when old equipment was being purged and it was given to him.
« Last Edit: September 02, 2018, 08:51:38 pm by Gregg »
Offline T3sl4co1l
    • Seven Transistor Labs
« Reply #2 on: September 02, 2018, 11:07:13 pm »
Quadrupole chamber made of hollow 'D' shapes, smells like a cyclotron thing.  Obvious problem, the lack of magnets (and lack of vacuum port too..?).
Seven Transistor Labs, LLC
Online coppercone2
« Reply #3 on: September 02, 2018, 11:20:59 pm »
that looks like an early model of the modern time machine
Online coppercone2
« Reply #4 on: September 02, 2018, 11:24:11 pm »
btw there is a wheystone bridge thingy drawn o nit
Offline T3sl4co1l
« Reply #5 on: September 02, 2018, 11:26:10 pm »
With a cam in the center.
Get it?  Cam bridge?!!?!....
http://www.richardsradios.co.uk/cambhistory.html
At least they had a sense of humor in their early days... :)
Online coppercone2
« Reply #6 on: September 02, 2018, 11:42:46 pm »
i thought it was a physic symbol
« Reply #7 on: September 02, 2018, 11:43:04 pm »
Tim,
Thanks for finding the link to the logo.  Now we at least know it isn't a schematic of the device.
Offline chris_leyson
  • Country: wales
« Reply #8 on: September 03, 2018, 12:03:14 am »
Apparently it's a quadrant electrometer but missing the quartz suspension fibre, mirror and small vane that sits inside the brass quadrants.
https://physicsmuseum.uq.edu.au/quadrant-electrometer
https://www.orau.org/ptp/collection/electrometers/Cambridgedolzaleck.htm
The following users thanked this post: Gregg
« Reply #9 on: September 03, 2018, 12:14:13 am »
Thanks Chris, your google-fu far exceeds mine.  I'll have to dig through the remains of the estate and look for the vane and mirror.  If I find it, I'll certainly post pictures.  Maybe my friend was going to make a vane, since he was an instrument maker.  One of the many projects he never got around to doing.
Offline Gyro
Re: Cambridge Instruments – Dolezalek Quadrant Electrometer
« Reply #10 on: September 03, 2018, 11:55:43 am »
Well done to Chris_Leyson, I'm surprised the rest of you guys didn't get it (cyclotron? :o), it should have been covered in middle school or high school Physics - electrostatics.
Yes it is a very nice Cambridge Instruments quadrant electrometer (I've been after one for my collection for ages!). It looks in very nice condition, apart from the missing vane and mirror and at least 100 years old. The lacquered brass-work looks really well preserved. The quadrants work as differential pairs, attracting and repelling the leaves of the vane.
Regarding the vane, it might actually still be there - @Gregg, you mentioned that the quadrants aren't quite aligned. ONE of pair of the quadrants should be able to pivot so that the vane can be inserted and removed, you might find that it is still inside. Unfortunately you are missing the tiny circular mirror and suspension fibre but you might be able to fabricate them (if you can find an old galvanometer suspension for instance), look around on the web for an old book on electrostatics. They would probably have been held together with shellac, the vane would be thin Aluminium foil in the shape of a fat flattened 2 vane propeller.
In terms of perservation, the lacquer on the brass is soluble in most solvents so soft dry cloth clean only. Those nice orange insulators need special care too - they are made of very high insulation resin - fused Sulphur was used at one time, so again, very gentle removal of dust with a soft cloth only. Dust attracts moisture, which gets through the lacquer and causes dark spots.
Thanks for sharing the detailed photos!
P.S. You can see a 'cheap' educational version here: https://www.ebay.co.uk/itm/A-RARE-ANTIQUE-PHILIP-HARRIS-QUADRANT-MIRROR-ELECTROMETER/323426129823?hash=item4b4db3179f:g:iRoAAOSwjrlayx1e
« Last Edit: September 03, 2018, 12:14:16 pm by Gyro »
Best Regards, Chris
Offline Gyro
« Reply #11 on: March 16, 2024, 02:10:38 pm »
I'm reviving this thread as, after years of searching, I'm now the proud owner of a Cambridge Dolezalek Quadrant Electrometer.
My example is a Cambridge Scientific Instruments Company Ltd model, dating from around 1906. Unlike the OP's it has it's inputs brought out on the underside and has a square window. I was also lucky that it came complete with mirror assembly and very luckily, when I opened up the quadrant, I found the original silvered paper vane which had become detached.
The company history is quite interesting, one of the founders was Charles Darwin's son. It was founded in the late 19th century, later getting into electrical measurement instruments, and introducing the quadrant electrometer in 1906. In 1919, the company merged (took over) with the Robert W Paul Instrument company to become the Cambridge and Paul Instrument Company, and then in 1924 changed it's name to the Cambridge Instruments Company, adding the familiar 'bridge' logo seen on the OP's example. Robert W Paul was an interesting character, in addition to his Scientific instruments company, founded in 1891 (I have one of his early High Sensitivity suspended coil Galvanometers, and also one of his Unipiviot ones), he was also the pioneer of cinematography in England, demonstrating moving picture projection in 1896 (around the same time as the Lumiere brothers in France) but after Edison. He had his own film studio.
I've attached some photos of the repaired Electrometer. I've re-fitted the vane, not easy because it is extremely fragile and had been torn from the mirror assembly. Whilst I have replaced the suspension, with a fine piece of glass fiber from a fibre-optic bundle (instead of the piece of cotton that saved the mirror from being lost), this still comes in at around 40um, which is still well short of the 4, 6, or 8um fused quartz suspensions that it originally came with. I've no idea where I would find one of those - I'm sure I have some single mode fibre somewhere, which has a core diameter of 7um which might get me close.
With the correct 6um suspension, these electrometers were capable of measuring below 0.1V potential differences across the quadrant inputs once the vane was charged to around 100V - it has a true differential floating inputs. I had always associated 'passive' electrometers with higher voltage measurements.
I've attached some photos, together with the relevant 1906 pamphlet - it's interesting to see that, at the time, their phone number was 'Cambridge 6'! Hopefully this is of interest to the OP and maybe others.
Best Regards, Chris
The following users thanked this post: Gregg, RoGeorge
Offline Kleinstein
« Reply #12 on: March 16, 2024, 02:39:25 pm »
The single mode fibers may have some 5-8 µm of core diameter, but ther total filter is much thicker, more like 100-200 µm.
The following users thanked this post: Gyro
Offline RoGeorge
« Reply #13 on: March 16, 2024, 02:41:16 pm »
In the pdf it says the instrument has a thin 0.004 mm diameter and 60 mm long quartz fibre.  :o
How do they made a quartz "wire" so thin and so long?  Is that quartz fibre brittle when bent?
Offline Kleinstein
« Reply #14 on: March 16, 2024, 04:01:39 pm »
Chances are one can make the extra thin quarz fibers just like the normal fibers for optics. So heat a piece if quarz and than pull a fiber. AFAIK the hotter the glass the thinner the fiber and also faster pulling can make it thinner.
The fiber would not be espeically sensitive and surprisingly strong. It would still be brittle and fail with no warning before it fails. With the thin diameter there is a good chance to be free of any defects to start with and thus quite strong.  Bending may allow a radius of possobly down to 100 diamaters or in the 0.5 - 1 mm range.
One might get relatively thin fibers from glass fiber insulation material - not the the same composition, but could be good enough an insulator.
Offline Gyro
« Reply #15 on: March 16, 2024, 05:22:40 pm »
Thanks Kleinstein. I remember heating and pulling glass tubes to make micro pipettes, as you say, the faster the pull, the thinner the result.
I don't know the relative tensile strength of glass and fused Quartz, I suspect that the latter is probably stronger, allowing smaller diameter. Even 40um fiber is surprisingly 'springy',  it's quite easy to sense when you are about to exceed it's bending limit, I have snapped them at as large as 10mm or so bend radius, which is surprising. Looking closely under the microscope, the old fiber bundle that I have looks to be surprisingly uncontrolled in fibre diameter, so I think it's going to be a matter of experimentation now I have the technique.
Spun glass loft insulation may well be a possible source, as long as I can get sufficient length, I considered silk too, but everything I've tried so far is woven from strands that are too short.
Cambridge also offered a fine (80um x 4um) Phosphor Bronze ribbon suspension for permanent, rather than brief, application of bias but I haven't found anything near small enough, and they had lower sensitivity anyway.
I've attached a few other non-pdf images from other (slightly later) catalogue pages.
« Last Edit: March 16, 2024, 05:28:36 pm by Gyro »
Best Regards, Chris
Offline RoGeorge
« Reply #16 on: March 16, 2024, 07:07:43 pm »
How sensitive is such an electrometer in practice?
For example, if I connect 2 metal plates at a 9V battery, and keep the plates separated at 9cm apart, between the plates it will be a DC electric field of about 1V/cm.  If I place a 3rd electrode in the air, between the 2 plates, will the electrometer be able to sense the field between the plates?  0.1V would correspond to moving the 3rd electrode with 1mm between the plates, will an electrometer be able to sense that?  (never had access to an electrometer, no idea what it can do in practice)
« Last Edit: March 16, 2024, 07:12:32 pm by RoGeorge »
« Reply #17 on: March 16, 2024, 07:14:55 pm »
Thanks Chris for your contribution to keeping ancient technology alive.  I haven't played with my electrometer in quite some while.
I did try to separate a single filament from some braided Dyneema fishing line with limited success (I couldn't separate long enough fibers).  Maybe I'll have to try different sizes and brands of line.  If I can get a filament thin enough, Dyneema seems to be almost perfect for the task.  Kevlar is a possible second choice.  Both will allow knots or loops for the ends whereas glass or quartz probably would need to be glued.  Maybe carbon fiber?
Any suggestions for making a new stem for the vane would be appreciated; presently I am using the thinnest aluminum hobby tube I could find and a tiny non magnetic stainless screw at the bottom to hold the vane I made.
Unfortunately I cannot see as well as I used to and my dexterity is also suffering from becoming older.
Offline Gyro
« Reply #18 on: March 16, 2024, 07:22:41 pm »
In theory, yes. In practice, it would come down to insulation resistances. The terminals and segment supports are made of "ambroid", which from what I've read is a heated and compressed form of natural Amber. This would have very high insulation resistance if clean. The electrometer quotes 170mm deflection with the scale at 1m distance and a 100V bias on the vane. So it would resolve much smaller than this, eg. 1.7mm/mV.
Without the correct thickness suspension, I can't say at the moment but I think it would have a pretty good try. Of course you also need to consider the quality insulation of the 3rd electrode between the other two and screen against other disturbing electric fields.
Best Regards, Chris
Offline Kleinstein
« Reply #19 on: March 16, 2024, 07:34:20 pm »
For the part holding the mirror, I don't see a need to go extremely low weight. The weight still loads the fiber, but especially if the fiber is a bit thicker this should not be that much of an issue.
Aluminum could be an issue in making good contact. So at least the lower part, up to the contact should be more like a hard more noble metal to make easy contact and with a small diameter (to have low friction when rotating). So more like brass, nickle or bronce wire and than only aluminum for the upper part holding the mirror and hook, that no longer needs to be round. I would go as far as considering some 0.2 mm Manganin (a kind of bronze) wire up to and including the hook, and glue the mirror to it.
Online coppercone2
« Reply #20 on: March 16, 2024, 07:41:52 pm »
can you measure capacitance portion of input impedance?
Offline Gyro
« Reply #21 on: March 16, 2024, 07:57:47 pm »
Thanks Gregg. Sorry the info is a case of better late than never, finding the stuff after acquiring mine has been a bit of a detective job. There are very few museum photos and most show battered and incomplete examples. The Smithsonian has quite a good archive of old Cambridge catalogues though, but as I indicated, nothing on the electrometer until 1906.
Thanks for the Dyneema reference, I'll take a look. Yes, kevlar and carbon fibres are possibilities, I don't know how fine they get. In the Carbon case you would need a permanent bias supply for the vane - this would need to be applied relative to both segment connections, which would be inconvenient in a high impedance situation. There is a terminal for this at the top of the center screw, initially I though the knob was just a mechanical stop, then I noticed the cross hole drilled near the top of the screw thread. With an insulating fibre, you only need to charge the vane periodically using the 3rd terminal with the rotating strip at the bottom.
My hands are getting shaky too, attaching the fibre to a couple of wire hooks with cyanocrylate adhesive proved a lot easier than trying to tie knots (not that this is an option with glass fibre) although I did need a magnifier. Presumably Cambridge would have used the same method, probably using shellac.
It sounds as if you had some success with the vane. Hopefully the photos give some idea of the original. As I mentioned, the 1906 model used a silvered paper vane (I don't know if it would have been Mercury amalgam silvered or chemically deposited from Silver Nitrate). I notice the 1908 version mentions an Aluminium vane, so thin foil would do the job. The rest of the assembly is just wire and shellac, the mirror is about 5mm x 0.5mm, front silvered (mine is a bit tarnished but usable). I saw somewhere that it is slightly concave, with a 2m focal length. Looking around, a small disc of reflective window film might make an acceptable substitute - it is available on ebay in 'sample' size.
« Last Edit: March 16, 2024, 08:53:47 pm by Gyro »
Best Regards, Chris
Offline Gyro
« Reply #22 on: March 16, 2024, 08:07:28 pm »
Tricky, it's going to be very low, and the vane is electrically isolated. One of the documents refers to the electrostatic capacity being "about 80cm", I've no idea how that would translate to modern units, maybe it is an equivalent measure to the distance between two air spaced plates [Edit: or spheres] of standard area?
« Last Edit: March 16, 2024, 08:10:39 pm by Gyro »
Best Regards, Chris
Offline Gyro
« Reply #23 on: March 16, 2024, 08:45:15 pm »
Just looking through my electrometer folder. Here are a couple of photos of the later (probably Cambridge instruments era) Aluminium vane assembly...
Edit: Looking at the images, the mirror must be thinner than 0.5mm.
« Last Edit: March 16, 2024, 08:55:46 pm by Gyro »
Best Regards, Chris
Offline Gyro
« Reply #24 on: March 31, 2024, 07:26:49 pm »
Kevlar to the rescue... well mostly.
I found some Kevlar sewing thread on ebay that, from the photo, had a nicely splayed end. When it arrived, I found that I was able to fairly easily untwist the thread and liberate individual fiber bundles and then fibers, which my micrometer tells me are 6um. Unlike cotton and silk thread, the individual Kevlar fibers are continuous (for more than the length I need anyway). I needed a stereo microscope and a few attempts to attach a single, nearly invisible, fiber to the hooks, the easiest way was to clamp the first hook, apply cyanoacrylate, and then wave a bundle of loose fibers until one of them caught. The second hook was harder as I was handling a single temperamental fiber.
I've attached a photo of the 40um glass fiber and the 6um Kevlar fiber assemblies for comparison. While the glass fiber is rigid enough to hold the two (30awg) hooks at 90' to each other, the Kevlar is completely 'flexible' but easily strong enough to support the mirror and vane assembly. After leaving overnight for any twist to relax, the mirror position is stable. The original Cambridge booklets indicates a mechanical time constant of 35 seconds and "practically dead beat" with the 6um Quartz thread, the Kevlar gives significantly greater than this, probably around 2-3 minutes, with no overshoot. The 40um glass fiber, by comparison, gives around 2 seconds, with significant oscillation.
I'm now able to see significant deflection when applying potentials across the quadrants (I haven't got my scale and light source set up yet). I've been unable to find much on the electrical characteristics of Kevlar fiber, other than it having a moisture content of around 3.5%. This probably makes it resistive in electrostatic terms, so it probably needs constant bias to be applied to the top terminal.
The search for true 6um quartz fiber continues, but in the background now.
Best Regards, Chris
The following users thanked this post: Gregg, zrq