I have been living in the dark ages a bit. On my bench, I have 3 Fluke 8050A non-autoranging multimeters, and the main handheld DMM I have been using has also been non-autoranging. I actually like selecting the range I want, and the readings are instant.
I have several autoranging DMM's but it struck me I never use them for measuiring current.
In my recent post on the Amprobe AM-270 multimeter, I was a bit shocked to see a 1.65V burden voltage at the top of the 500mA range. The reason was of course is that they use one resistor for the 500uA and 5000uA ranges, and one for the 50mA and 500mA ranges.
So if a meter needs at least 100mV to get full range at a 50mA current , then it will have a 1V burden voltage at 500mA.
Are all the autoranging DMM's bad like this? Are there any good ones? It seems to me that they should just have the switch select all current ranges explicitly, and forget about the fiction of autoranging for current.
If I buy a Fluke 87 and measure 400mA on the 400mA range, I will have a burden voltage of 720mV.
If I buy a $10 multimeter with a manual ranging switch, I will probably be able to measure the same current with a burden voltage of less then 200mV.
Go figure!
Richard.