That multimeter had a major input overvoltage, it takes a few kV to arc over to the adjacent part and wreck that resistor. It looks like it arced to the missing outside resistor.
If it carbon tracked the PC board then the damage is too severe to repair IMHO. Must be a microwave oven job.
Because the parts are old and have a safety function, they are not easy to substitute.
Fluke p/n 831594 RES, CERM, 909kΩ 1% 2W 100ppm
I recognize that as obsolete IRC GS-3 series, now Welwyn/TT Electronics GS-3 thick film metal glaze resistor, 1,000V surge rated, 5.7mm D 13.1mm L.
This is the brown resistor you have that is damaged from the arc.
Older 87's used (blue) Vishay Dale
FP69 3W MOX 150ppm flameproof.
Fluke p/n 822015 THERMISTOR, RECT, POS, 1.5kΩ 30%
The PTC thermistor needs to be 1,000V rated. All I am aware of is Vishay PTCCL07H100VBE but it has no UL approval for some reason.
Second contender is G.E. now Amphenol YS4020 but it is 1.1kΩ.
Fluke p/n 1275713 RESISTOR, CERMET COMPOSITION, 1kΩ, +-10%, 1W, -1300+-300PPM
Original KOA Speer PCF-1C102K, possible substitute Ohmite OX102KE.
Fluke p/n 1277360 THERMISTOR, POSITIVE, 1.1kΩ, +-20%, COATED, RADIAL
Original G.E. YS3961, possible substitute Amphenol YS4020.
The missing resistor is either 3.5kΩ or 1kΩ surge-rated so check your service manual.
Not sure but 1st gen 87
Mr. Modemhead repair the resistors are switched in position.
edit: added the 87V 1k surge resistor and PTC for people doing repairs on others.