I'm repairing a failed Tektronix AFG-3022C with an interesting problem at power supply.
The power supply is an Artesyn 7001318-J000 module.
At first, it was not turning on.
Suddenly, it started to work (and I checked the voltages with the poor diagrams of the service manual and was everything ok)
But in this condition, the equipment was resetting if it suffered a minimal movement.
After a disassembly and reassembly, it started to work without resetting, but now it stopped to recognize the ASICs (was working even with the resetting issue).
I checked the PS again and discovered that the voltages are altered (5V to 7,1V, +/-15V to +/-18V; the 4.3V keeps almost the same). Weird detail: at the parts list section of the manual, the power supply is described as having these altered voltages as features.
Does anybody have a clue about what is happening here?