I’ve been reading up on the principles of 2465A calibration and trying to weigh up the strategy going forwards.
To dip my toe in the water, I tried the easiest calibration routine namely CAL05 (hours of operation and number of power-ups), resetting both to zero. The machine appeared to save this data, since it appeared on the next power-up (incrementing the number of power-ups by one as expected) so from that I’m assuming the RAM itself and its battery is OK, at least for the moment. The checksum error still occurs, so CAL05 didn’t rewrite it successfully – from that I’m assuming that the checksum isn’t a single global value that’s automatically regenerated after every successful partial CALxx (although some online comments suggested this might be the case).
There is currently no trace (just display and scale factors) even using beam find. I embarked on the start of CAL08 (CRT adjustments) and the Grid Bias test displayed the single spot as it should, which I could adjust with the Grid Bias pot. After that, however, I had to abandon CAL08 due to lack of test equipment as I haven’t even got a simple function generator to put a sine wave on the screen as needed for the later focus adjustments. There’s still no trace, and my assumption here is that the calibration constants have been so badly damaged that the machine can’t even display a simple horizontal sweep. There seem to be plenty of adjustments in the various CAL routines that, if not correctly done, could kill the trace entirely. So if things are this bad, the checksum problem is entirely secondary and will clear once the calibration is done correctly.
So my plan is to:
1) Replace the Keeper battery – although the voltage is OK, it’s the original one so clearly it should go, and anyway I’ve got nothing more to lose from disconnecting the battery at this point
2) Acquire enough test equipment to get through the calibration somehow. I’m not averse to spending money as I need a function generator and a frequency counter anyway, but it’s a shame that the 2465’s requirements are rather more specialised than anything I’m ever likely to need again.
What worries me more about the world of calibration isn’t the complexity as such, but all the online talk about ‘hidden’ procedures not in the manual, and the possibility of getting permanently stuck in a calibration loop because the machine thinks something is out of range and won’t exit/store cleanly. I’ve already noticed that the CAL08 tests don’t come up in the order advertised in the manual. My serial number is 131499, and I’m using the updated manual pages dated 6/5/87 for serial numbers B016276 and above, but I wonder if there was a later firmware change and manual update that I haven’t got?
If I run into calibration problems, which seems more than likely, I’d be grateful for the advice on forum etiquette please: should I carry on here, start a new thread or add to the main 2465 Teardown thread?