Hi, all-
I'm in the process of troubleshooting a Tektronix 485 oscilloscope, and I currently have the transformer and inverter boards on my bench to isolate a fault (which doesn't appear to be present on these boards after testing). These can be found on pages 152 to 159 of the
service manual. Anyway, they were a bit of a PITA to get out of the chassis, so while I have the boards out I've been testing any capacitors that regularly see high voltage. Most of these are ceramic, which AFAIK are touted for their superior ESR. So far, almost every single one has tested open circuit at ~100 Hz, ~1 kHz, and ~100 kHz. I should mention that I'm testing in-circuit, but the test signal applied by my meter is around 225 mVpp. Also, I could see in-circuit testing leading to artificially low values due to parallel impedances, but I don't see how it would give an artificially high value. I've used the same meter to find a dried up electrolytic filter cap in a 475, and have tested a number of resistors for calibration, and known-good and known-bad caps for confirmation that it functions as expected.
I find it hard to believe that every single one of these have failed, especially since the 'scope powers up and works
mostly okay. I.e., the major functions all operate, but with some eccentricities that need resolving. So anyway, what gives? Is this operator error? Do high voltage ceramics have unusually high ESR?
Thanks!
Edit: I should also mention that I've found one film cap that measures open, in addition to one that measures sub 1 ohm on the inverter board. The plot thickens...