I have an old Tektronix 2213 that sat in a box on a shelf for a decade, then was flown across the country in a suitcase in checked luggage (but still in the box and well padded!). When I turned it on it hissed and there was no image. I took it apart and found a leaked capacitor (photos below, click them for full size). It's C971, which according to the
Service Manual filters the +8.6V low voltage supply.
I couldn't find an 840uF capacitor but I bought this 820uF
ELXY250ELL821MK20S from Digikey. I replaced it, and the scope works!
Questions:- Should I be concerned about the 20uF mismatch? I figured it's only 2% off, and the original cap was a Sangamo 301EN841U012B2 which I believe had a 20% tolerance. Should I replace C975 (or any of the related filtering caps) to match?
- It took more force and (and heat) than expected to desolder the dead cap. They sure made things more durable back in those days! Any tips in case I find myself needing to replace other components? Note I was using solder braid since I don't have a pump.
- My scope was stored in a plastic bag, tied with a knot, inside a box. When I opened it up there was a strong smell, which I assume was a mix of "Eau Du Old Tandy computer" and leaked electrolyte. After removing the cap and letting it all air out for a day, the smell isn't as bad. Should I be worried about it (e.g. toxicity?) and do I need to do any other cleanup (e.g. wipe down all the surfaces, probe sheathing, knobs, etc. with isopropyl alcohol)?
- I measured the Probe Adjust point with my multimeter (nothing fancy, just a run of the mill Radio Shack 22-168A). It's labeled ~1KHz and ~500mV peak-to-peak, but my meter reported 883kHz and 273mV AC (not sure if that's RMS, or if it matters for a squarewave). Does the discrepancy indicate other problems?
- How's my trace focus and calibration look?
I realize I shouldn't have a long wire(/antenna) hooked up on the end of my probe like that; just wanted to show the MM for reference.
- After using the scope for a while, I noticed the waveform occasionally jumps up or down by several volts, off the screen. After a while it goes back. Am I doing something wrong, or might this be a symptom of other issues?
- Has anyone found good quality scans of the schematics? Particularly #9, which shows the power supply.
- I haven't performed any measurements of the test point voltages inside the case. Given the high voltages inside, I'm a little hesitant to do too much with the cover removed, particularly with the high-voltage shield removed. Any advice for doing so safely? (I've read some people keep one hand behind their back to reduce the likelihood of shorting anything through their heart).
- Anything else I should check/do?