I just got a good deal on a 485. one channel will only trigger on .1-.5 changing volts/div to 1-5V or 5-50mV it will not trigger. this is with a 10V sine wave with a 10x probe with the scope set at 1Mohm.
If I decrease the voltage to 5V and use the 50ohm button and a 1X probe It will trigger on all settings.  Any ideas.
Re: Tek 485 trigger
« Reply #1 on: September 23, 2014, 11:03:00 pm »
That changes input to 1 V sine wave but a V/div setting of less than .1 will have the waveform much larger than the display.
This is never a good idea with a CRO. The Vertical amp is overloaded.
The 50 ohm input is attenuating the signal to manageable levels.
Do you have the same result with 5 V sine and 10 x with 1 M input?
Get the Tek troubleshooting guide:
http://www.sphere.bc.ca/test/tek-parts/troubleshooting-scopes.pdf
Have you a Service manual?
Plenty of links to find one here:
https://www.eevblog.com/forum/repair/repair-documents-and-links-sticky-me-please-mods/msg477909/#msg477909
Re: Tek 485 trigger
« Reply #2 on: September 23, 2014, 11:26:07 pm »
5V sine ,10x probe setting,  1V per division, 1Mohm setting. No trigger.    5V sine, 10x probe setting,  .5V per division setting, 1Mohm setting. Trigger and wave form. Channel 2 triggers with no issues on all settings.
Offline liquibyte
Re: Tek 485 trigger
« Reply #4 on: September 24, 2014, 06:23:55 pm »
I hooked an ohm meter to the output the working channel shows the correct 1Mohm through all divisions. Non working channel shows 1Mohm on divisions 5V-.1V  then 0ohms in the 5-50mV.  50ohm setting shows 50ohm through all divisions on both channels and is working properly.
Re: Tek 485 trigger
« Reply #5 on: September 24, 2014, 07:00:02 pm »
Sounds like the Ch 1 input attenuators may have been damaged by an input overvoltage event.
« Last Edit: September 24, 2014, 07:45:33 pm by tautech »
Re: Tek 485 trigger
« Reply #6 on: September 24, 2014, 07:47:26 pm »
I have swapped the attenuators from the working channel to the non working channel and the other way around and it made no difference.
Re: Tek 485 trigger
« Reply #7 on: September 25, 2014, 07:32:04 pm »
Figured it out it was a broken contact on the hi-z board. the plastic had cracked allowing the contact to float around.
Re: Tek 485 trigger
« Reply #8 on: September 25, 2014, 07:59:30 pm »
Good find.  :-+
Can you specify the connector number from the schematics so others with a similar problem know which one to suspect/check.
Re: Tek 485 trigger
« Reply #9 on: September 25, 2014, 11:48:04 pm »
Tektronix part# 131-1031-00  on the back of the hi-z board.