Alcohol cleaning didn't work. I think it's a mechanical problem.
I don't have an exact Tek push button switch, but I do have a low-profile, short travel push button switch from my parts collection. The legs do not align, the Tek switch has a wider pitch... so bent the legs of my switch slightly, to go into the original holes. Made sure the switch is level and soldered it.
Next problem, the push button caps don't work with my switch. It doesn't have the right "profile" for the switch cap to snap in. So I had to break the legs of the switch caps so the switch caps will sit directly on top of plunger, and be able to actuate the switch.
Screwed together the front panel switch and pot board, making sure I can still hear/feel the clicking of all the switches while I align the switch caps with the panel. At this point, feeling good that the replacement switch is working.
Installed front panel assembly back to the scope, connected the 2 ribbon cables and powered it up.... pressed the SLOPE switch and now it works, and changes from (+) to (-) and vice versa.
Now, the real goal.... go into Diagnostics mode. Press Delta-V, Delta-T, and Slope.... YES! Scope goes into Diagnostics mode. Did the All Test, and everything passes test.
Now I can continue following the rest of the service manual Calibration procedures.
Since all TESTS ran successfully, I'm thinking maybe the scope just badly needs calibration?
So that's my next to do item.