Author Topic: TEK 2245A Start-Up Problem  (Read 756 times)
Offline JdaleHTopic starter
« on: November 12, 2023, 06:02:18 pm »
When I press the on/off switch, most of the time the unit will not start up.  I hear an ongoing click sound (relay?).  I can usually keep pushing the button and the unit will turn on.   Most aggravating.  Can it be fixed?
Offline Jeff eelcr
Re: TEK 2245A Start-Up Problem
« Reply #1 on: November 12, 2023, 10:39:14 pm »
Yes, sounds like a capacitor problem in the power supply as a guess but would require troubleshooting to the problem.
Start with service manual and multimeter at a minimum, capacitance checker may help.
Jeff
Offline Jeff eelcr
« Reply #2 on: November 14, 2023, 01:11:07 am »
Did you mean Tektronix 2445A?
Jeff
Offline Vovk_Z
  • Country: ua
« Reply #3 on: November 14, 2023, 05:05:01 am »
Quite often that's a power supply problem (one or several dried electrolyte capacitors). The most efficient  way to troubleshoot it when you have in-circuit ESR meter.
Offline BlownUpCapacitor
  • I kinda suck at the art of electronics tbh
« Reply #4 on: November 14, 2023, 08:31:27 pm »
Go to PDF page 263 on the service manual here: https://w140.com/tekwiki/images/d/d1/070-7672-00.pdf
disconnect power supply output wires to completely isolate the PSU from the scope. If the 2245A uses PCB jumpers instead like traditional 22xx series scopes, desolder those. Measure voltages with a DMM and use a constant current load to measure the tripping current and make sure it matches up with the specs in the service manual.
If the PSU trips under loaded, trouble shoot the PSU. If the PSU works perfectly fine, trouble shoot the rest of the scope by disconnecting power to some parts of the scope and see what makes the PSU trip. Isolate the problem and troubleshoot the suspect board/board area.
Go to the Maintenance section of the manual and the sub section "Troubleshooting Techniques". Read that and then start your troubleshooting journey.
Hehe, spooked my friends with an exploding electrolytic capacitor the other day 😁.