Got the burst measurements uploaded from memory using the GPIB interface.
The commands have to be issues in a specific manner when doing this manually.
Set instrument range eg 0.1v
Clear the memory history
Set up the instrument to trigger in burst mode - set no. samples required.
Set the GPIB to ++auto 2
send TRI to start the burst mode.
Set to ++auto3
Send Du (DUmp memory contents)
Send ++Read
memory contents are loaded into the serial monitor screen.
Graph below.
Sample rate is 1Khz, the sine wave period is 30ms or 33hz
I did the same on Range 1 and range 10, but the signal is very "digitised" so not very helpful but is a similar but 'blocky shape'.
Would be interesting to see what your instrument shows Grizewald, performing the same test
![Wink ;)](https://www.eevblog.com/forum/Smileys/default/wink.gif)
So that is something to think about.
Next I'll swap out the ICL7650 and do the same test.
![Solartron7061](https://www.eevblog.com/forum/index.php?action=dlattach;topic=232622.0;attach=983250;image)