Hello Rich
today I opened my CRTU which I guess has the same PS (PULS SN250) and 'probably' the same connectors at the control side. I believe at least the former is true, so here's what I did:
1) check voltages at PSU connector with unit turned OFF (but powered)
2) check voltages at PSU connector with unit turned ON
3) identify possible "power on" lines at PSU connector
3) turn off and disconnect unit, find those 'candidate' lines at control side (X12 connector in my unit)
See attached pictures and PDF. Pins pictured in green, output some voltage even in the OFF state.
So, I'd say B1, C1 and D1 are pretty good candidates, especially B1 given that the ON logic could be negative.
If i were you I'd measure the voltage at B1 with the PSU disconnected from the equipment, and if you get some voltage, drive it to zero with a
10K to 100Kohm resistor
Hi MasterTech,
thanks for the response and the work you put into this.
It would be nice to have the voltages on the pins while it is working and while it is in standby.
Looks like the same connector alright, may even be the same motherboard too.
Here is a sketch of the connector I did last week. At the lower left and at the far lower right are the pins I had marked "to check" up to the front CPU module. Much like you suggested.
As you can see B1 C1 and D1 are on the list. Also are A2 A3 D3 A24 B24 and D24 that need to be traced on the MB.
Also is a pix of the PSU label with the type number. Looks like it is the same PSU. Not a surprise considering the lineage of the two units.
There is also another separate connector that goes to the first plug-in. It looks like and old pc hard drive connector.
I have the PSU powered on the bench with the standby 12V line active.I'll try a 10k resistor to drive the B1 line to zero as you suggested and try other combinations.
Cheers and thanks. This is an on-going saga......
![Wink ;)](https://www.eevblog.com/forum/Smileys/default/wink.gif)
UPDATE:Just measured the voltages. PSU on the bench (not connected to the motherboard but under power):
(pin#, voltage)
a1 10.7V
b1 11.1
c1 0
d1 0
a2 0.1
a3 0
d3 0.5
a24 1.0
b24 0
d24 12.4
UPDATE2:
Drove a1 and b1 to near ground with a current limit resistor (about 1k). No effect. They are high internal resistance lines as 10k brought them to about 5V.
Also d24 is a solid 12V. Probably the 12V standby.
I used two voltmeters to look at the two major set of pins that look like high current output, most likely the system 5V and such. I would have noticed if the PSU came to life.
So does not look like a1 or b1 are negative logic to start it. Perhaps some other pin are positive logic and require the 12V to start the PSU?
Would be nice to take a full readings on a good unit while in "standby" and then "on" and see what are the difference on these lines indeed.