Author Topic: repairs of HP 8594A spectrum analyser  (Read 1899 times)
Offline kenrameshTopic starter
  • Country: in
« on: April 17, 2018, 09:13:49 am »
hello all ,
I have been facing problem with HP 8594A spectrum analyser.
1.As the instrument is switched on , FREQ UNCAL is displayed which is a permanent fixture.
2.When i try  to calibrate using CAL OUT the result is  SIGNAL NOT FOUND & SRQ110.
3. The trace line is slow in movement , data entry response also appears to be slow.
I couldn,t find service manual for 8594 series , but tried using 8590 series one.
As per that manual either A7 analog pcb, A16 processor , counter lock , or YTO may be faulty but not able to zero-on by using this manual.
I observed the following:
1.on A7 analog interface board Q1 is getting warm
  but if  cable running to YTO removed from the PCB, heat is normal.
2. FREQ UNCAL disappears & trace line moves fast when Counterlock board flat cable is removed
waiting for ur help  Pls suggest how do proceed.
Thanx in advance
Locked due to duplicate thread: https://www.eevblog.com/forum/repair/help-needed-to-repair-hp-spectrum-analyser/msg1494742/#msg1494742
« Last Edit: April 26, 2018, 06:19:33 am by gnif »