I received my knock-off push-pull plugs (FGG.2B.305.CLAD99Z) from AliExpress, and realized that it's built for much thicker cable (9.2 to 9.9 mm). The sales listing didn't specify which collet it would come with and its photo showed a collet for thinner cable. I'm going to try to buy a thinner collet.
Speaking of which.... what cable to use? My most recent poor-man's idea to to use SuperSpeed USB cable:
Signal | Wire | connector 2B.305 pin |
Lo | Black | 3 |
Hi | Red | 4 |
Guard | Drain+shield, isolated from connector housing | 1 |
Lo-sense,High-sense | shielded twisted pair, drain to lo-sense | 2,5 |
(floating) | shielded twisted pair + unshielded pair | |
My thought is to put high and low next to each other to try to have them have a more similar temperature.
I know the wire insulation may have issues with dielectric absorption (and leakage), but I think that the shielding should be reasonable, and the 20 or 22 AWG red and black should be good for current-carrying. Use of copper wire (in physically small connectors) should reduce thermal voltages.
The LEMO connectors have a "test voltage" of 1.60 kVrms body-to-shell, and suggest a max operating voltage of 1/3 or 583 Vrms. Multiply by sqrt(2) to get a 816 VDC (or so) limit.
The proper Fischer connector has a contact to body test voltage of 2.4 kVDC. Dividing by three, we get 800 V, so about the same.
However, knowing that USB cable is probably specified at max 500 V, I should keep it below that, though it'd likely work at much higher.
(Capacitors and optocouplers have been replaced, boards partially cleaned, will fully clean them once I know everything is working, still waiting for the new input sockets. Also, I noticed that in TEST10 mode, a relay switches for every measurement. I'm guessing that's normal.).