Thanks for the explanation James. I didn't realize that resistor was such a critical part.
So it got me thinking
I have never chose "50 Ohm" inputs on my TDS420 and I imagine it is used when the input is something other than a probe. So I own one 36" coax cable terminated with BNC connectors. The cable is stamped Belden 8259 RG58 A/U. I want to test the four channels of the scope individually to see if the attenuator resistors are still good.
I set the Siglent 1032X to 50 Ohm output and to 7.562 kHz at 255.1 mVrms for the following tests:
In the first test, I set the scope to "50 Ohm" inputs and found that all four channels hit the frequency but the amplitude is slightly low:
For the second test, I switched the oscilloscope to "1 mOhm" inputs and attached a 50 Ohm terminator to the scope's inputs and all four channels hit the frequency but the amplitude turned out to be more accurate:
I imagine when using the 50 Ohm input setting that the type of cable, and its length, needs to be matched to the scope?
*Update* Check this out. So I sent the signal from the Siglent 1032X to channel 2 of an HP 35660A. The DSA seems to back-up what the Tektronix scope read for amplitude. The HP measured 254.38 mVrms and the Tek measured and rounded to 254.4 mVrms!
This sort of tells me that the TDS420 50 Ohm attenuator input resistors are fine and, maybe, that the cable and/or the terminator needs slight adjustment.
*Update* I checked the resistance of the BNC output jacks of the Siglent SDG-1032X with two different multimeters today while the Siglent was set to "50 Ohm" output.
Both multimeters read 50.4 Ohms for channel one and 50.5 Ohms for channel two of the Siglent.
As for the 50 Ohm terminator, it is made by Amp and it measures spot-on at 50 Ohms.