Hi George
After I posted last time, in response to your post I began to doubt my findings, so I stripped out resistor board and checked the readings again, I still got the same sort of results. So I lifted one end of each and low and behold, they were spot on. I then removed the AC transformer and popped the board into a ultrasonic cleaner and gave it a couple of goes. Once it had all dried out nicely, I resoldered the lifted ends back on the board and checked values, again spot on. Then I reattached the transformer and ribbon cables, checked values again, and they were back to the original values
Prior to the above I had stripped the meter down and disconnected everything as in the manual and removed the top switch board. Ribbon cables all look to be in good condition, the top switch board seems to be OK, contacts look nice and clean, no pitting etc and good contact is made with the moving contacts. Bottom contacts also inspected and the ribbon cable connections checked and all are good, a real head scratcher this one.
Today, following your post, I rechecked the total resistance per range via the plus and minus terminals on the DC ranges and using my Brymen BM867 (goes upto 50M) so I had to use a cheap manual switcher to read the 3,000V reading as it goes upto just over 60M. Any way I obtained the following readings.
DC Volts Resistance.
3 60.3K
10 201.5K
30 666K
100 2.06M
300 5.03M
600 7.78M
1000 10.8M
3000 58.8M
I did notice that the readings were creeping upwards if I left the meters connected, just like a capacitor being slowly charged?
I then got out my other 2 Model 8's, not to sure which marks they are but both have the 3,000 range as opposed to more normal 2,500v One is the normal domestic specification and the other is a NATO issue one and weighs a bleeding ton. These are the reading obtained from them in the same fashion as the Mk5.
Normal one
3 30.1K
10 200.8K
30 603K
100 2M
300 6M
600 11.99M
1000 19.88M
3000 40.5M
NATO one
3 60.3K
10 200.3K
30 603K
100 2M
300 5.9M
600 Not supported on this meter
1000 19.4M
3000 39.98M
Neither of these two meters exhibit the slow upwards creepage that the Mk5 does and the capacitors on the Mk5 are minimal values and should not be in circuit as the AC side is switched out of circuit? Could the ribbon cable be acting as a capacitor I wonder?
To answer your other question about has a D cell leaked at some point, I think the answer to that is yes some evidence of slight corrosion on the plus terminal but zero sign of anything inside the meter or the battery compartment itself.